[evlatests] Test for 10-second dropout problem
Rick Perley
rperley at nrao.edu
Wed Sep 2 10:21:16 EDT 2015
Vivek ran an hour-long script yesterday, which was designed to
characterize the '10-second' problem.
The idea was that if the effect is due to the change in
tuning/configuration between X-band pointing and x-band observing which
uses a different correlator/tuning setup, we should try multiple
observations of this transition, compared to a transition using the same
tuning/configuration between pointing and observing. Thus, we generated
a script which did the following cycle eight times, observing the
super-strong calibrator 3C273:
X-band Ref pnting (two-SPW mode)
X-band observing (wideband)
X-band Ref Pnting (two-spw mode)
X-band observing (two spw mode)
The effect I saw in the October data from last year was *not*
reproduced. However, a similar, and much more comprehensive effect was
seen:
1) The transition from two-SPW pointing to two-SPW observing at the
same frequency setup showed no amplitude dropouts of any kind.
2) The transition from two-SPW pointing to multi-SPW observing
showed complete absence of coherent visibilities at the beginning of
each scan, with the following characteristics:
a) SPWs 2 through 8, and 9 through 16 showed 10 seconds (exactly)
of complete noise at the beginning of each scan -- every scan. Both
polarizations, and all antennas are affected equally. The return to full
amplitude is immediate -- no transition.
b) SPW 1 did not show this effect at all -- looks just like the
two-SPW situation.
c) SPW 9 showed a 'partial effect' -- the first ~ 4 seconds had
reduced amplitude.
In my October calibrator data, I did not check to see if the
dropout behavior in SPWs 1 and 9 were different than the others. I'll do
that later this morning.