[evlatests] Switched Power Review -- A Truly Great Result
Rick Perley
rperley at nrao.edu
Wed Feb 5 18:05:00 EST 2014
It has been a few months since I last reviewed the status of our
switched power calibration system, for all antennas and all bands. The
'stress test' run this morning provided an opportunity for a systematic
review. Only ea14 was out of the array. All other provided good PSum
and PDif data at all bands.
Bottom Line: Only one antenna, at one band, in one polarization,
gave a non-usable switched power:
ea16, in RCP, (IFs A and B) had zero-mean (just noise)
PDif.
All other switched power data looked great!
A secondary, but still important, and definitely intriguing result
is this:
ea06 and ea16, at S-band, did *not* give the oscillatory
PDif values that I saw in every other S-band database that I've ever
looked at (and that is over 100 by now). So what's different here? One
thing is that the requantizer gains were set -- but Ken and I agree this
should have no effect on the switched power. But maybe it does?
Another possibility is that ea06 and ea16 were both 'being nice' at the
time -- the stress test is only at S-band for 3 minutes, and the
oscillatory phenomenon is know to change its nature from time to time.
Another possibility is that the location of the observation (in the far
north, at declination +78) is free of whatever is bothering ea06 and
ea16. But I've reviewed two long runs from the ARCADE project --
declination +57, and these showed the phenomenon at full strength.
More investigation is warranted, which Ken and I will do.