[evlatests] Standing Waves and X-band
Rick Perley
rperley at nrao.edu
Fri Apr 26 16:45:15 EDT 2013
The short test done last night provides some useful information
into bandpasses. I used only a single subband (to keep data volume
down), but this is sufficient ...
All bands show a wobble in the bandpasses (both amplitude and
phase) of typically 23 MHz period -- this is clearly due to a standing
wave set up between the receiver input and the subreflector: the length
corresponding to a 23 MHz period is 6.5 meters, and the distance between
the subreflector and the secondary phase center is 6.6 meters.
The reflections at X-band are more pronounced than at other
bands. The typical pk-pk power amplitude of the standing wave is ~0.5
dB, which I think corresponds to a ~-15 dB return loss. Even more
intriguing, for most antennas, the standing wave pattern is quite a bit
larger at the high frequency end of the bandpass -- typically twice the
amplitude at 11.5 GHz than at 8.5 GHz.
One antenna is particularly bad: ea11, on the RCP side (IF 'B', at
11.5 GHz), has a pk-pk amplitude of nearly 3dB in the standing wave (and
about 20 degrees in phase). The corresonding return loss is -8 dB. The
pattern is barely visible in the LCP side (at the same frequency).
More information about the evlatests