[evlatests] Wobbles at X-band
Rick Perley
rperley at nrao.edu
Tue Jun 22 11:22:31 EDT 2010
The 'Cygnus A' experiment run last week has clearly revealed the
presence of 'wobbles' at X-band.
These are seen only seven baselines (out of 325 in the array at the
time). The characteristics are the same as has been reported earlier
with the prototype correlator. As noted there:
a) Periods are typically a few seconds.
b) On the baselines which show both amplitude and phase wobble, the
phase period is half that in amplitude.
c) The effect is typically evident for an hour or two. During this
interval, the period of the phenomenon changes, reaching a maximum in
the middle. There is good evidence that the wobble phenomenon is
*always present*, but is only visible when its period is long enough
that it is not averaged out by the integration period.
d) The effect is on both IF pairs, in both polarizations, with equal
amplitude.
e) The frequency of the wobble is proportional to the observing
frequency.
f) The phenomenon does not close -- it is not possible to factor out
the effect by antenna.
g) There is no relationship of the presence of the wobble, or its
frequency, to baseline or source geometry.
The baselines on which this is seen are given below, with some notes:
4 x 20: (W1 x N5) amplitude and phase. Seen in the middle of the run.
4 x 28: (W1 x N8) Phase only. Seen in the middle of the run.
2 x 10: (E2 x N3) Amplitude and phase. Seen in the beginning of
the run.
2 x 14: (E2 x E5) Phase only. Seen in the beginning of the run.
5 x 25: (W8 x N2) Amplitude and Phase, seen at the end of the run.
8 x 24: (N1 x W5) Phase only, seen at the end of the run.
5 x 15: (W8 x W6) Phase only, seen at the end of the run.
I have not yet checked the C-band or L-band data (taken at the same
time) for the presence of this phenomenon.
More information about the evlatests
mailing list