[evlatests] Effects of T304 attenuator settings on bandpass shapes
Michael Rupen
mrupen at nrao.edu
Mon Jun 29 15:05:07 EDT 2009
The "0,0" levels for the three antennas in this test were as follows:
Input attn Output attn
ea01 7 18
ea18 9 14
ea25 12 15
Rick, would you send around a PDF/PS of your BP plots? I'd like to see
whether the 7->8 and 15->16 changes had any greater effect, as Jim (?)
thought they might.
Michael
> Earlier WIDAR tests have shown that changing the T304's attenuator
> levels causes changes in the bandpass amplitude shape by ~1%. To
> determine which attenuator is responsible, whether particular settings
> have a more dramatic effect, and to otherwise illuminate the
> characteristics of the problem, Ken and Michael ran a specific
> experiment. In this, they stepped through five different input and
> five output attenuator settings, spending one minute on each pair. This
> was done on three antennas (1, 18, 25), the others were fixed. The
> observations were of 3C84 in one continuous observation. The attenuator
> pairs which were sampled were (in order): (0,0), (-2,0), (-1,0), (0,0),
> (1,0), (2,0), (0,0), (0,-2), (0,-1), (0,0), (0,1), (0,2), (0,0). All
> values are in dB.
>
> The data were calibrated and the bandpasses determined for each
> antenna, for each of the paired settings listed above. The AIPS program
> BPLOT was used to show the *differential* results. (This means that the
> average of all bandpasses is made and subtracted. The results displayed
> are the fractional deviations from the average).
>
> Results: Not as simple as had been hoped.
>
> 1) The three antennas which had the attenuators changed clearly
> show larger changes in their bandpass shapes. The maximum deviations
> are ~2.5%. No commonality nor trends are seen in the bandpass changes.
> 2) The seven antennas with fixed attenuators showed much smaller
> deviations, much less than 1% for nearly all. Changes that are seen
> are gradual, indicating some physical change over time. Antenna 19 is
> by far the worst amongst these, with a 1.5% change over 13 minutes. The
> amplitude pattern has a ~100 MHz periodicity, indicating a standing wave
> set up over ~1.5 meter length.
> 3). Two of the changing antennas -- 18 and 25 -- showed by far the
> biggest deviations (>2%) with the setting (0,1) and (0,2) -- the output
> attenuator seems to have the biggest effect. This is also true for
> antenna 1, but the amplitude of the bandpass change for this antenna is
> very much less (~1%).
> 4) For the three antennas with attenuators that changed, the four
> (0,0) settings all showed different bandpasses. These changes are
> typical for the temporal variation of the seven antennas with unchanging
> attenuators.
>
> It thus appears likely that we have two simultaneous effects
> confusing the issue: A temporal change, operating over timescales of
> tens of minutes, and an attenuator-depending effect. Both are of
> similar magnitudes -- up to 2% in bandpass amplitude.
>
> All the above applies to the amplitudes only. I'll determine the
> phase effects after running FRING to remove residual delay effects (this
> should keep my computer busy over the lunch hour ...)
>
>
> _______________________________________________
> evlatests mailing list
> evlatests at listmgr.cv.nrao.edu
> http://listmgr.cv.nrao.edu/mailman/listinfo/evlatests
>
More information about the evlatests
mailing list