[evlatests] Effects of T304 attenuator settings on bandpass shapes
Rick Perley
rperley at nrao.edu
Mon Jun 29 13:39:31 EDT 2009
Earlier WIDAR tests have shown that changing the T304's attenuator
levels causes changes in the bandpass amplitude shape by ~1%. To
determine which attenuator is responsible, whether particular settings
have a more dramatic effect, and to otherwise illuminate the
characteristics of the problem, Ken and Michael ran a specific
experiment. In this, they stepped through five different input and
five output attenuator settings, spending one minute on each pair. This
was done on three antennas (1, 18, 25), the others were fixed. The
observations were of 3C84 in one continuous observation. The attenuator
pairs which were sampled were (in order): (0,0), (-2,0), (-1,0), (0,0),
(1,0), (2,0), (0,0), (0,-2), (0,-1), (0,0), (0,1), (0,2), (0,0). All
values are in dB.
The data were calibrated and the bandpasses determined for each
antenna, for each of the paired settings listed above. The AIPS program
BPLOT was used to show the *differential* results. (This means that the
average of all bandpasses is made and subtracted. The results displayed
are the fractional deviations from the average).
Results: Not as simple as had been hoped.
1) The three antennas which had the attenuators changed clearly
show larger changes in their bandpass shapes. The maximum deviations
are ~2.5%. No commonality nor trends are seen in the bandpass changes.
2) The seven antennas with fixed attenuators showed much smaller
deviations, much less than 1% for nearly all. Changes that are seen
are gradual, indicating some physical change over time. Antenna 19 is
by far the worst amongst these, with a 1.5% change over 13 minutes. The
amplitude pattern has a ~100 MHz periodicity, indicating a standing wave
set up over ~1.5 meter length.
3). Two of the changing antennas -- 18 and 25 -- showed by far the
biggest deviations (>2%) with the setting (0,1) and (0,2) -- the output
attenuator seems to have the biggest effect. This is also true for
antenna 1, but the amplitude of the bandpass change for this antenna is
very much less (~1%).
4) For the three antennas with attenuators that changed, the four
(0,0) settings all showed different bandpasses. These changes are
typical for the temporal variation of the seven antennas with unchanging
attenuators.
It thus appears likely that we have two simultaneous effects
confusing the issue: A temporal change, operating over timescales of
tens of minutes, and an attenuator-depending effect. Both are of
similar magnitudes -- up to 2% in bandpass amplitude.
All the above applies to the amplitudes only. I'll determine the
phase effects after running FRING to remove residual delay effects (this
should keep my computer busy over the lunch hour ...)
More information about the evlatests
mailing list