Michael has observed the Orion core in three 'chunks', each 1.024
GHz wide. I have done the basic calibration for the first of these --
23630 to 25655 MHz.
All antennas fringe nicely, although antenna 27 is (as seems to be
usual, for reasons unknown) weak by a factor of at least 10.
There are a very small number of 'zero records' -- 0.018% of the
visibility data are zeroed. These are easily identified and removed. I
see no evidence for any significant number of reduced visibilities that
cannot be ascribed to antennas off source, or other antenna-based effects.
I used POSSM to check a few of the raw spectra, looking for
'displaced' subbands. All spectra were continuous in amplitude and
phase across subbands.
Basic calibration was done: FRING to find the delays. SETJY/GETJY
to establish the flux densities. CALIB to calibrate the subband gains,
BPASS to remove the bandpass shapes.
The calibrated spectra are -- with a few exceptions -- smooth in
amplitude and phase across subband boundaries. The only exceptions are
between subband 1 and 2, where some baselines show a small apparent step
in amplitude. No explanation is offered, other than the obvious one
that this was introduced by the calibration, since no step occurred in
the raw data. Subband 1 is peculiar -- there is clearly very high
aliasing of external signal on its lower end -- perhaps some other
effects are making antenna-based calibration ineffective in this
subband. (That is, the data don't close).
At least 9 spectral lines are visible in short (couple of minute)
vector integrations.
I tried making an image of the continuum and of one of the lines --
but the results of this were puzzling. It would be good if somebody who
has knowledge of the expected structure could get involved now.
Rick