[evlatests] C-Band observations of 0217, Thursday
Rick Perley
rperley at nrao.edu
Sun Aug 30 14:53:18 EDT 2009
John Benson managed to convert the C-band tests to UVFITS on
Friday. The results from the test are:
In general, the data are of very good quality, with no serious,
obvious problems. The problems noted are all the same as reported
Friday for X-band.
1) All 12 antennas fringed strongly in both RCP and LCP.
2) There are a significant number of zero amplitude/phase records --
roughly 0.1% of the data. These are easy to remove.
3) The delays are all exactly the same as they are at X-band, (to
with 1 nsec or so), except the signs are reversed. Note that the very
large values and differences between the antennas in LCP are present at
C-band also. Although this is a very minor problem (since FRING finds
and removes these with high accuracy), it would be good to fix up the
on-line delays, so the phase slopes are smaller. (Makes editing the
data easier, if nothing else).
4) Bandpass solutions are good, except (as at X-band):
Antenna 3, LCP, subbands 3 and 4 have a strong slope in
amplitude. (Subbands 1 and 2 look normal, but the sensitivity is much
reduced in these two, so I removed all four).
Antenna 8, LCP, all subbands, as a pure, 1-cycle sinusoid (in
frequency) bandpass.
I removed the LCP subbands for both antennas from the database.
5) Antenna gains show beautifully smooth phase variations, exactly
as expected for the baseline lengths and frequency at this time of
year. The amplitude solutions show the same curious instabilities that
are visible at X-band, to wit:
- All antennas show non-sinusoidal variations, typically +/-
2% in amplitude (or 4% in power). Some antennas reach 6% at times. All
four IFs in both polarization show *exactly* the same variations. All
antennas have independent patterns. Since the data are in no way
normalized, we are seeing here changes in attenuation or gain. At this
frequency, I think we can discount opacity effects, which leaves us with
amplifier gain variations (or perhaps, but much less likely, pointing
problems).
6) Following a 2-point amplitude and gain calibration, noise
histograms were plotted. All antennas in both polarizations gave
beautiful gaussian histograms, with 1sigma widths of 0.42 Jy. A BOE
(back-of-the-envelope) estimation is that this should be 0.33 Jy -- we
are somewhat noisier than we should be. However, this observation was
on a very strong point source -- true noise estimates must be made in
blank fields. I suggest this be done, when convenient.
7) Clipping discrepant visibilities removed only 0.3% of all
visibilities (including the edge channels, were gain variations on
filter edges will surely cause some fluctuations) -- this is very good.
8) Imaging showed that again that the noise does not go down with
sqrt(BW), after about 32 central channels are included. With both
polarizations, all four subbands, and 150 central channels, the noise
ends up a factor of 2 to 3 above that expected from the histogram
width. The images show the single background source I'm aware of, but
also some 'lumpy' stuff nearer the point source, or similar amplitude.
Closure corrections reduce the 'lumps', but have little effect on the
noise. The edges of the images are dominated by patchy sinusoids (the
patches are symmetric about the center, but are located well past the
first null of the beam -- they are clearly an imaging artefact). I'm
not at all sure about what to expect with so little data, and such large
fields -- only 16 minutes of data are in this test.