[evlatests] A strange problem at 12.5 MHz BW -- in continuum
Rick Perley
rperley at nrao.edu
Sun Jul 27 19:03:11 EDT 2008
In an effort to better understand sensitivity issues, I've taken
data at X-band, in continuum, of a known calibrator and an adjoining
blank field, at each BW from 50 MHz through 0.78 MHz. All of the
expected effects are seen -- I'll show these at the Monday test meeting.
But one thing quite unexpected was also seen -- a notable
degradation in sensitivity, affecting BW = 12.5 MHz only.
The loss of sensitivity -- which is seen *both* in noise histograms,
*and* in the correlation coefficients, affects all correlations equally
-- hence, both IFs (which were tuned to different frequencies). And it
affects equally both EVLA and VLA antennas.
More curiously -- the loss in sensitivity is greatest for antennas
at the end of the east arm (about 20%), and least for antennas at the
ends of the other arms (less than 5%). The gains determined for the
antennas at this bandwidth are also variable. (All other gains, are
rock-solid).
This smells like external RFI, originating near the end of the east
arm -- except this doesn't explain why both IFs are equally affected, or
why no other bandwidth is affected.
So is it possible that this is some internally-generated signal,
which manifests itself in the spatial sense noted above due to
differential fringe rates? The 'u' baseline coordinate was near zero
for the baselines comprising east-arm antennas.