[evlatests] Consequence of IAT clock error
Claire Chandler
cchandle at nrao.edu
Tue Jan 8 11:43:17 EST 2008
Are there tests that should be done during start up that could check for
this in the future?
Claire
On Tue, 8 Jan 2008, Rick Perley wrote:
> The clock error which was found and corrected yesterday appears to
> have resulted in the following:
>
> a) A loss of ~ 25% in sensitivity on EVLA to VLA baselines
> b) Closure errors of up to 100% on EVLA to EVLA baselines.
>
> Hence, data taken from the time the clock error began (probably
> Thursday during maintenance time -- but this has yet to be established)
> until its correction (during software time yesterday) are severely
> compromised. Users should flag their EVLA to EVLA baselines. EVLA to
> VLA baselines do not show any closure problems, but are reduced in
> sensitivity. VLA to VLA baselines are unaffected.
>
> I believe that the new 'L350 1PPS' device, soon to be implemented,
> should eliminate this kind of mis-timing. Testers should note that by
> far the easiest way to detect this problem is to set the closure report
> level (within CALIB) at 50% or so. (I had discontinued looking at
> closure myself because of the large number of moderate level (10 -- 20%)
> errors from the bandpass mismatch problem was cluttering up the screens).