[evlatests] stress test results of Dec. 9
Emmanuel Momjian
emomjian at nrao.edu
Wed Dec 12 20:45:56 EST 2007
The following are the results of the Dec. 9 stress test.
VLA 2: X and K bands, IF C only: the correlator coefficient values show
a phenomenon that has a cycle time of 30 seconds: half of the cycle has
good values, and the other half has values that go up and down by +/- 10
to 15% with respect to the good values seen in the first half. A similar
behavior is also seen in C-band with a shorter cycle time (about 20
sec). The source observed in C band is different from the source
observed in X and K bands. Nothing unusual is seen in the phases when
the above is occurring.
Specific Bands:
K-band:
-EVLA 26: IFs B & D have poor sensitivity compared to IFs A & C. IF B
is worse by a factor of 1.5 and IF D is worse by a factor of 2.3 with
respect to IFs A & C. However, the system temperature values are similar
in all four IFs.
X-band:
- VLA 5 and 20: phase jump after the P-band observation (this is a
VLA to VLA phase jump, as the reference antenna for calibration was VLA 9).
C-band:
-EVLA 17, 21, 23, 24: short shallow drops in amplitude.
-EVLA 26: has no fringes in IFs A, B, and D, and hardly any in IF C
(signal to noise is so low that solutions were impossible to get).
-EVLA 18: IF A has significantly low correlator coefficients (by a
factor of 10). Its Tsys is 3000 K compared to IFs C & D that have 35-40K.
-EVLA 18: IF B has significantly low correlator coefficients (by a
factor of 42). Its Tsys is 1.5 K only compared to IFs C & D that have
35-40K.
L-band:
-EVLA 11, 17, 24, and 25: short shallow drops in amplitude.
-EVLA 17: IF A has higher Tsys values compared to other IFs by a
factor of 1.5.
-EVLA 21: IFs A and B have higher Tsys values (factor of 2 compared to
IFs C and D).
-EVLA 26: IF C lots of drops in the correlator coefficient values,
Tsys shows variations between 30 and 140 K.
IF B Tsys shows variations between 30 and 200 K.
IF D has correlator coefficient values lower than those in
IF A by a factor of 2, but Tsys is the same and stable around 33.5 K in
both these IFs.
P-band:
-VLA 7: weak or no fringes in all IFs.
-EVLA 13: weak fringes (the correlator coefficients are at about 50%)
in all IFs.
-EVLA 18: No fringes in all IFs, and Tsys values are several 1000 K
(between 4000 and 25000 K).
-EVLA 25: weak or no fringes most of the time in all IFs.
-EVLA 26: no fringes in IFs B & D only.