[evlatests] L-band switched power results from the band-switch test
Rick Perley
rperley at nrao.edu
Mon Nov 12 18:43:17 EST 2018
35 observations of J0359+5057 were made, each one minute long, in 7
blocks of 5 each. In each block, the frequency was cycled back and
forth through the other 7 cassegrain bands. System flags only were applied.
1) For the SPWs without RFI, the switched power was extremely
stable, with a single exception (see below). No irregularities or signs
of multiple levels, or instabilities were found.
2) For one antenna-IF only, non-proper behavior was seen: ea22R has
an ~12% modulation, of period a couple of minutes. This is seen in
multiple spectral windows (I have not checked them all).
I calibrated these data with, and without the switched power applied.
1) For the 'raw' data (no switched power applied):
- slow gain variations, with timescales of tens of minutes, and
amplitudes of up to 8% (in power) were common. ea12 and ea16R were the
worst. For most of the others, variations were in the 2 -- 4% range.
2) For the data with switched power applied:
- All gain variations noted above are removed, such that residual
variations are at the 2% level, or less, with a single exception: ea22R
has a ~12% (power) modulation, clearly caused by the variations in the
switched power (PDif) values noted above. This proves that the
fluctuations seen in the antenna-IF do not reflect gain changes.
More information about the evlatests