hi all,
A while back I sent some results from a test that compared S-band
switched power measurements done with the P-band cals switching vs off
(see evlatests email on 2016/06/17). That previous test was done using
a single 128-MHz-wide spectral window centered near 3 GHz. I've since
repeated the same test using the full receiver BW at both L and S bands.
Results for a few antennas are attached. The top panel of the plots
shows the ratio of the P-band cal signal to the L or S cal, measured in
two different ways: 1) the usual switched power table values; 2) WIDAR
autocorrelations, which provide better frequency resolution. These two
methods seem to give consistent results. The bottom panel shows the raw
"on-off" autocorrelations with either the P-band cals on (red) or with
the in-band (L,S) cals on (blue).
The main takeaway from this is that this P-band cal leakage has lots of
structure on ~10-100 MHz frequency scales. My previous ordering of
which antennas were "good" or "bad" mostly reflects whether or not they
happen to have a big spike near 3 GHz. For example ea03 (the worst one
last time) would have looked great if I'd done the original test near
2.5 GHz instead. When viewed across the full 1--4 GHz, all the antennas
have similar features somewhere in the band.
This may not be a big problem for current observing practices, where
flux calibrators are observed frequently and the switched power is
mainly used as a kind of intermediate scaling. I think this effect
would be an issue if we were to move towards a scheme where the switched
power is used to set flux scales. Also while the average power level
increase across >100 MHz bands is not too big, over small freq ranges
this may be large enough to noticeably affect Tsys.
In general this seems like a messy thing to be happening and I think we
should either filter the P-band cal signal appropriately, or else just
turn off the P-band cals during non-P projects. I realize VLITE is
always observing P-band, but it's not clear to me whether or not they
actually make use of the switched power signal; would be good to get a
definitive answer to that.
Cheers,
Paul
-------------- next part --------------
A non-text attachment was scrubbed...
Name: P_to_LS_leakage.ea03.png
Type: image/png
Size: 91869 bytes
Desc: not available
URL: <http://listmgr.nrao.edu/pipermail/evlatests/attachments/20160812/846d625e/attachment-0003.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: P_to_LS_leakage.ea13.png
Type: image/png
Size: 88621 bytes
Desc: not available
URL: <http://listmgr.nrao.edu/pipermail/evlatests/attachments/20160812/846d625e/attachment-0004.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: P_to_LS_leakage.ea18.png
Type: image/png
Size: 78450 bytes
Desc: not available
URL: <http://listmgr.nrao.edu/pipermail/evlatests/attachments/20160812/846d625e/attachment-0005.png>
More information about the evlatests
mailing list