[evlatests] 10-second tuning failures at X-band
Vivek Dhawan
vdhawan at nrao.edu
Mon Aug 31 10:35:05 EDT 2015
Sounds lke L302's failing to lock for 10 sec, a fairly common transient
effect on 1% of the 224 units. But why it is only at X band I cannot explain
- I assume you changed bands many times after the reference pointing?
On Mon, August 31, 2015 08:26, Rick Perley wrote:
| The calibrator survey run, taken last October, has revealed the
| continued presence of short-lived tuning failures. The characteristics
| are quite simple:
|
| 1) No fringes at the beginning of a scan, lasting exactly 10
| seconds. After this interval, full-valued fringes abruptly occur.
| Hence, this is not a pointing problem.
|
| 2) Roughly 20% of scans are affected, nearly always a single
| antenna, occasionally two. This seems consistent with a random
| process. There was no clear tendency for any particular antenna to fail
| more often.
|
| 3) The four IFs failed independently, with equally likelihood --
| failures are always on a single IF.
|
| 4) So far as I can determine, this problem only occurs at X-band.
| (I have checked P, L, S, and C, so far).
|
| 5) The on-line flagging system (unsurprisingly) does not flag these
| failures.
|
| These X-band observations followed X-band referenced pointing,
| which utilized a different tuning and correlator configuration.
|
| There was also a single incidence (out of 182 scans) where the
| *last* 10 seconds abruptly lost fringe power.
|
|