The calibrator survey run, taken last October, has revealed the
continued presence of short-lived tuning failures. The characteristics
are quite simple:
1) No fringes at the beginning of a scan, lasting exactly 10
seconds. After this interval, full-valued fringes abruptly occur.
Hence, this is not a pointing problem.
2) Roughly 20% of scans are affected, nearly always a single
antenna, occasionally two. This seems consistent with a random
process. There was no clear tendency for any particular antenna to fail
more often.
3) The four IFs failed independently, with equally likelihood --
failures are always on a single IF.
4) So far as I can determine, this problem only occurs at X-band.
(I have checked P, L, S, and C, so far).
5) The on-line flagging system (unsurprisingly) does not flag these
failures.
These X-band observations followed X-band referenced pointing,
which utilized a different tuning and correlator configuration.
There was also a single incidence (out of 182 scans) where the
*last* 10 seconds abruptly lost fringe power.