[evlatests] P-band status update: 15APR2014
Paul Harden
pharden at nrao.edu
Wed Apr 16 16:02:16 EDT 2014
On 4/16/2014 12:30 PM, Huib Intema (NRAO) wrote:
> Note that the
> following antenna/polarization pairs deviate from Paul's values:
> - EA02 RCP: 15K instead of 14K
Receiver SN06 in EA02 was repaired with hot/cold Tcal tests done March
10, 2014. Tcal values were 15K LCP and 14K RCP (Trx=75K and 69K LCP/RCP).
At these temperatures (non-cooled receivers), I can only measure the
Y-factor and noise CAL power (on/off delta) with about 0.1dB accuracy
with several seconds integration time, which equates to 2-3K accuracy in
deriving Tcal. Thus, I'm actually quite happy to see Tcals verified
within 1-2K.
> - EA22 RCP: 17K instead of 19K
EA22 has rcvr SN08, which was rebuilt and set to the lower Tcal values
on Jan. 23, 2014. Final Tcals were 14K LCP and 17K RCP. I don't know
where the 19K came from, probably my error, but will correct the website
and Tcal table immediately.
Range of hot/cold load test values at P-band are:
Y-factor: 3.9-4.2, 4.0 typ. (for noise figure 0.8-1.0 dB, 0.9dB typ.)
Pc-Pcc: 0.3 dB max. (cold difference between CAL on/off)
Measurement accuracy: 0.1 dB (+/- 0.05 dB)
Trx and Tcal power is measured at 350 MHz (mid-band) through a 25 MHz
wide filter. Measurements done from 200-450 MHz shows Trx and Tcal is
quite consistent (within 1-2K) across the pass band.
Paul