During the Holiday break we lost Weather Station Data around the
time
when we powered down the array for main power work. It was
found at the
time that we had lost a +5V power supply in the Weather Box
which was
replaced. However, this did not restore the information flow
into the
Monitoring system. Further troubleshooting by the DCS group kept
pointing to the CMP Computer, but it was felt that I needed to
confirm
that thinking.
Yesterday (1/9/2012) Terry Cotter and myself went to the site in
order
to determine if the LO/IF and M&C section of the Master Rack had
been
affected by efforts to remove the old VLA D-Racks and unused
sections of
the Master Rack. I determined that the timing signal (QQ) for
the SLC
and System Buffer was there and Weather Station Data was present
at the
Data Tap connected to the System Buffer. However, I had to
replace the
Dataset EPROM in order to restore the full Weather Station Data
flow.
At this point I looked at the Data Tap again for monitor points
from the
Weather Station (Power Supplies/Sacred +10V) that I would
recognized and
they were there.
For those of you who need to recall or don't know the fundamental
operation of of the old M&C system here is briefly what takes place.
The System Buffer at a time based upon QQ requests information
from its
Datsets (Weather = Dataset 0) and stores in into a serial memory.
At a different time but also based upon QQ the SLC requests the
stored
information from the System Buffer and echoes it back thus
displaying on
the Data Tap. At the same time, the SLC will transfer the
message using
three parallel transfers into the CMP for each message.
Simply put, seeing Data Tap data means that the System Buffer to
Dataset
link and the SLC to System Buffer link are operational.
I then tried the second SLC and still no success.
My conclusion is that the issue is either at the CMP or between
the SLC
and CMP.
At this point, I don't know what to look for in order to move
forward.
Finally, I have a very vague recollection that we needed for the
alternate Computer Port of the SLC to be plugged into the
Modcomp (even
powered down) for proper signal termination for the CMP to start
operations or to maintain operations. Remember this is vague and I
could have it wrong, but if this was stripped out then maybe
this is it.
-wayne-
--