Answers, below:
Michael Rupen wrote:
>> 1) Global problem: ea17 in LCP was providing extremely noisy data.
>> The PDif values vary over a ratio of nearly 2:1 at all bands, and in
>> both IFs B and D. The same variations are seen in PSum, but they are
>> proportionally much smaller -- so the derived Tsys also varies by nearly
>> a factor of two. The character of the variations is pure noise -- there
>> is no trend on any timescale.
>
>
> I assume this is NOT present in the visibilities?
You assume correctly.
>
>
>> 2) Band-dependent problems:
>>
>> a) ea28 in LCP at L-band. The system works well for about half the
>> time. For the other half, the values are many times higher, with
>> enormous variations in power. The system stays in each state for
>> typically an hour or two, then reverts to the other state. *** This
>> situation has never changed since the switched power system was
>> installed on this antenna***.
>
> Again, I assume this is NOT present in the visibilities?
>
Again, correctly.
>
> Have you played any games with applying the switched power from one
> subband to another? This might help avoid the RFI problems you
> mention. Or does out-of-band RFI affect *all* subbands equally? (which
> would be surprising)
>
This is my standard procedure, and works well. There is a scale
offset in the visibilities, but that is a small price to pay for getting
reliable gain correction. We need to investigate the WIDAR option of
placement of the non-correlating subband into a 'safe' RFI zone. This
is a major issue for L and S bands, especially for those experiments
which are not utilizing 'wide-band' mode.
> -- Michael