Rick Perley wrote:
> A quick review of the reported system temperatures reveals a few
> problems, mostly with the tabulated Tcal values.
>
> Q-Band: Tsys is claimed to be 20K on the A and C IFs. (We wish!).
> The problem is in the Tcal values, which are listed as 1K. (Curiously,
> the Tcal values for the BD side are find).
What antenna? What frequency?
>
> X-band: Tsys is between 750 and 1000 K on ea05. These look like
> the truth, based on system noise.
>
> U-band: Tsys is claimed to be 10K for ea12. The problem is with
> Tcal, which is listed as 1.0 K.
Obsolete Tcals replaced.
> Tsys for ea21 is between 50 and 60 -- looks like the
> truth. This is two times too high.
>
> S-band: Lots of problems here.
> ea07 has a (probably real) Tsys of 100 to 250 K.
Tcals are around 3, which is higher than most receivers
> ea15 has Tcals which are clearly 3 times higher than
> they should be (except for IF 'A', which is 300 times too high).
The Tcals in the database are around 1.5 to 2 on both IFs, which look
pretty reasonable. I don't know what is going on here. Possible
cal switching problem?
> Antennas 16, 18, 21, 23, and 26 all have listed Tcal =
> 1.0. I guess these are newly installed receivers, and nobody has
> updated the Tcals from their default numbers.
Don't quite know how this happened - I suspect they were previously
entered without the leading zero on the receiver serial number. Now
fixed.
> ea22 has a claimed Tsys of 10K. Here the problem is
> that the Tcal values are listed as 0.75 K or so -- clearly a factor of 2
> to 3 too low.
Obsolete Tcals replace with newer values.
> ea08 has a Tsys of about 50 on the BD side.
>
> L-band: ea07 has a (real) Tsys of 75K.
> ea18, on IFs A and C have Tcal values of 1.0 Seems
> unlikely, given the derived Tsys = 20K.
Fixed.
>
>