[evlatests] Bandpass Stability -- 3bit vs. 8bit.
Rick Perley
rperley at nrao.edu
Thu Nov 17 10:27:02 EST 2011
I reported yesterday that the bandpass stability from the 3-8 bit
tests was 'disappointing'. I've now had a chance to review this a
little more closely.
The bandpass stability is indeed disappointing -- but it has nothing
to do with the 3-bit samplers.
A bandpass solution was made on 3C84 (30 Jy) once every 8 minutes.
The duration of the 3C84 scan was 2 minutes, so the SNR is excellent.
Hanning smoothing was applied to prevent 'ringing' and improve SNR in
the vicinity of the resonance features.
The AIPS program BPLOT is very useful in showing differential
temporal variations. Plotting the three paths for the frequency span
common to all three (15.0 -- 15.896 GHz) shows the solutions are:
1) Nearly identical for all three paths. They are especially
identical (!) for the A1C1 and A2C2 paths. The 8-bit B0D0 path varies a
little in detail, but shows all the same major features. From this we
can deduce that the path division within the T304, and the subsequent
digital processing make very little difference to the bandpasses.
2) There are distressingly large temporal variations over the 2.5
hour of the experiment. How large? By 1 to 2% in amplitude (or up to
4% in spectral power), on frequency scales varying from a ~10 MHz to 150
MHz. Some antennas are especially bad: ea10 is easily the worst.
3) However -- there is hope. All these changes are gradual over the
observing window -- it's clear that an hourly calibration, or perhaps
twice hourly, will nicely remove these changes.
In important question is the origin of these variations. The
observation was made during a period of very large temperature range --
but also a period of rapid elevation rise. Which -- if either -- is
responsible?
I have plots, for interested eyes ...
More information about the evlatests