[evlatests] X-Band Sensitivity -- Superb!!!
Jim Jackson
jjackson at nrao.edu
Wed Nov 9 19:40:17 EST 2011
Rick,
Thanks for the encouraging message. We like to hear news like that.
One question: Would it be possible (or meaningful) to get these same
efficiency figures from the six antennas with the Teledyne 3-bit
digitizers? It might be interesting to see if or how bad the
problems with the 3-bit digitizers are actually affecting the bottom
line. Are we even in a stable enough condition to do this with the
3-bit digitizers yet?
Cheers,
Jim
At 09:23 AM 11/9/2011, you wrote:
> As part of the upcoming 'path to completion' review, I'm executing
>some tests to determine the EVLA's basic characteristics.
> Last night, under ideal conditions, I did a 'sensitivity' test:
>Using old-style 'OSRO' mode (two subbands, 128 MHz width, 2 MHz channel
>resolution), tuned to reasonably placed frequencies in each of the eight
>bands. I observed 3C147 to establish the amplitude scale, then to a
>blank field 2 degrees away to measure the noise. Referenced pointing
>was applied. The usual basic calibration was done.
>
> Here I report only on x-band -- the others will follow as time
>allows today.
>
> All antennas -- except ea05 -- which were in the array fringed as
>expected. (antennas 19 and 20 are out).
>
> From the naturally-weighted noise maps, we measure the SEFD directly
>using the map rms, using all working antennas.
>
> 8300 MHz: SEFD = 270 Jy
> 11300 MHz: SEFD = 250 Jy.
>
> We can get a good measure of the system temperatures and antenna
>efficiencies by using the switched power measures, after correction by
>the calibration from 3C147. (The 'raw' Tsys values use the measured
>Tcal provided from engineering. Any errors in these are measured by
>calibration against a source of known flux density, since the observed
>visibilities have been converted to Jy by the provided Tsys values).
>
> Adjusting the measured Tsys values by the gain factors gives the
>average Tsys: (these are eyeball averages -- but the spread is
>impressively small, so they should be pretty decent estimates).
>
> 8300 MHz: Tsys ~ 28 K
> 11300 MHz: Tsys ~ 26K
>
> Finally, we can then measure the efficiencies by using the SEFD values:
>
> effic. = 5.62*Tsys/SEFD (where 5.62 comes from the antenna's
>collecting area and Boltzmann's constant).
>
> 8300 MHz: Effic ~ 0.56
> 8700 MHz: Effic ~ 0.57
>
> I guesstimate the accuracy of these values as a few percent. A
>better analysis will follow, in due time (but well before the
>presentations to our august visitors ...)
>
> Bottom line: The sensitivity is considerable better than the
>Project Book specifications (SEFD = 340 Jy at 10 GHz). As the
>efficiencies are exactly what we expect (and they should be!!!), it's
>the outstanding receivers which are allowing us to boast (just a bit!).
>
> Congratulations to all!
>
>