[evlatests] Austin 1050 L-Band cycling problem
Rudy Latasa
rlatasa at nrao.edu
Mon Jun 6 16:43:38 EDT 2011
Summary of L-Band 1050 cycling problem antenna 26 versus antenna 6
One 1050 temperatures cold and stable
One 1050 temperatures cycling up to 35K on the 15 K stage and 150K on
the 2nd stage.
After several months of cooling down and warming up the L-Band
receivers on antenna 6 and 26, the temperature cycling problem with the
new 1050 refrigerator has been narrowed down to the 15K (1st stage) or
50K (2nd stage displacer). Below I have listed all of the possibilities
that were suspected and looked at as the problem causing the temperature
cycling associated with the new L-Band receivers.
1. Receiver Thermal short
2. Temperature differential (Night & Day)
3. Compressor Output Pressures and flow
4. Gas line restriction
5. Refrigerator Drive Motor Power
6. Refrigerator displacer clearance, timing, seal dimple and valve leakage
After discussing our findings with engineers at Austin Scientific
(Oxford Instruments) on 06/06/2011, Austin Scientific requested that we
send these two refrigerators back to Austin Texas, so that they can
evaluate the two refrigerator driver/displacer units to determine if the
problem is in the 15K or 50K displacer.
Antenna 6 and 26 will be warmed up on 06/07/2011, so that these
refrigerators can be pulled and shipped back to the manufacture for
evaluation. Two new 1050 refrigerators will be installed in these two
receivers and cool back down. I expect the temperature cycling to continue.