[evlatests] Calibrator Polarizations from 1 to 50 GHz
Rick Perley
rperley at nrao.edu
Mon Feb 28 17:01:31 EST 2011
The 'flux densities' run has given us the calibrator source
polarizations, spanning the entire range between 1 and 50 GHz. I give
below, at selected frequencies, the *fractional* polarizations, in
percent, and the *true position angle* for the four primary
calibrators. All position angles are referenced to 3C286 = 33 degrees.
Freq. 3C48 3C138 3C147 3C286
---------------------------------------------------------------------------------
1050 .35 25 5.58 -14 <.05 xxx 8.6
1450 .50 140 7.50 -11 <.05 xxx 9.45
1640 .65 -5 8.45 -10 <.04 xxx
9.86
1950 .86 +30 9.00 -10 <.04 xxxx 10.12
2450 1.40 60 10.45 -9 <.05 xxx 10.50
2950 2.00 80 10.75 -10 <.05 xxx 10.75
3250 2.50 88 10.02 -10 <.04 xxx 10.90
3750 3.20 96 RFI <.04 xxx 11.10
4500 3.85 105 10.0 -11 .12 -100 11.25
5000 4.20 108 10.4 -11 .32 0 11.40
6500 5.20 112 9.80 -12 .25 -65 11.62
7250 5.25 113 10.0 -12 .57 -39 11.70
8100 5.35 -65 10.4 -11 .70 -25 11.95
8800 5.40 -63 10.1 -9 .77 -12 11.9
12800 6.00 -63 8.4 -8 2.2 42 11.9
13700 6.10 -63 7.9 -8 2.4 47 11.9
14600 6.40 -64 7.7 -9 2.7 52 12.1
15500 6.45 -65 7.4 -10 2.9 58 12.2
18100 6.90 -68 6.7 -14 3.4 65 12.5
19000 7.1 -69 6.5 -15 3.5 66 12.5
22400 7.7 -72 6.7 -18 3.8 73 12.6
23300 7.8 -72 6.6 -19 3.8 74 12.6
36500 7.4 -80 6.6 -27 4.4 82 13.1
43500 7.5 -88 6.5 -30 5.2 83 13.2
---------------------------------------------------------------------------------
The interesting new behavior is that of 3C138 -- the fractional
polarizations at high frequencies have dropped considerably, and the
p.a. has changed over the past 10 years: In 2000, the fractional
polarization was uniformly 11.5% from 4.5 through 45 GHz, and the
position angle slowly changed from -11 degrees at 5 GHz to -16 at 45
GHz. This source has doubled its high frequency flux density over this
period of time, so we are clearly seeing the effects of a new 'blob'
with a different polarization p.a.
More information about the evlatests
mailing list