[evlatests] Re-formatted version of Rick's message
Michael Rupen
mrupen at nrao.edu
Fri Sep 10 12:07:12 EDT 2010
Following Ken's lead, here's Rick's message again, with the first
table re-formatted for easy reading:
>From rperley at nrao.edu Fri Sep 10 09:30:56 2010
Date: Fri, 10 Sep 2010 09:29:09 -0600
From: Rick Perley <rperley at nrao.edu>
To: evlatests at aoc.nrao.edu
Subject: [evlatests] Some clarity on 3 bit sampler issues ...
Michael and Ken ran a test yesterday afternoon in which:
1) The power levels to the samplers were set at appropriate levels,
as judged by state counts, and
2) The stationboard requantizer gains were adjusted to optimum levels.
This was done, separately, for both 3-bit and 8-bit paths.
The observations were taken in the 'narrowband' mode, with 1 subband
at 7056 MHz (thankfully free of RFI). Observations were made of 3C286
and a nearby piece of blank sky.
The results have considerably clarified what is happening.
A) RMS noise levels.
I give the two 'noise matrices', with 8-bit on the left, and 3-bit
on the right, and make some comments below.
8-Bit 3-Bit
------------------------------------------------------
| 12 15 22 28 | 12 15 22 28
------------------------------------------------------
12 | X .15 .15 .16 | X .22 .45 .25
15 | .15 X .13 .14 | .17 X .33 .18
22 | .15 .13 X .15 | .18 .16 X .37
28 | .16 .14 .14 X | .19 .17 .16 X
------------------------------------------------------
Two observations from this:
1) In the LCP side, we have good uniformity in the noise values for
both 8-bit and 3-bit paths. But, the 3-bit path is about 15 to 20%
noisier.
2) On the RCP side for 3-bit, we have -- as usual -- values for
22A which are far too high. But I argue also that 12A is also
too high, so there is only one 'representative' value in that
matrix -- 15 x 28, whose noise value of 0.18 is right in line
with the LCP correlators. This interpretation is strongly
supported by other evidence, given below. Bottom Line --
I believe the RCP-LCP asymmetry that we have been focusing
on for the past week is a *** red herring***.
B) Gains -- as determined from cross-correlation power.
If all antennas are identical, and all system parameters set
equally, we should see the same cross-power amplitudes, and hence the
same calibration gains. Significant deviations from uniformity then
give us useful information.
Below I list the calibration gains (in power, not amplitude,
units). Note that a low number means the antenna's correlated power
(that is, the power from the source which is provided to the correlator)
is high.
8-Bit 3-Bit
-----------------------------
12R 36 69
12L 32 44
15R 24 36
15L 27 37
22R 28 144
22L 27 34
28R 30 44
28L 32 37
----------------------------
Note that 3-Bit gains are generally higher -- indicating that the
power from 3C286 to the corrrelator is lower -- and that 12R and 22R are
much worse than any others.
C) Autocorrelation Spectra.
Back in VLA days, autocorrelation spectra were never utilized, as
they had a 'bad rep' -- they gave weird results. But in principle,
these should give us useful information -- when properly calibrated,
they should tell us the antenna SEFD.
It appears that this is what the WIDAR correlator's autocorrelation
spectra are giving us. Below I give the calibrated autocorrelation
spectral amplitudes, in Jy.
8-Bit 3-Bit
------------------------------
12R 280 500
12L 260 330
15R 220 270
15L 220 260
22R 250 1200
22L 230 260
28R 250 350
28L 250 310
-----------------------------
A few spot checks shows a ***perfect*** correlation between the
geometric means of the products of autocorrelation amplitudes (above)
and the noise values in the matrix at the top. In other words, the
(easily obtained) calibrated autocorrelation amplitudes are an excellent
predictor of system sensitivity (as they should be). We seem to have
gained an even better way to determine system sensitivity than the old
(VLA) route of utilizing normalized correlation coefficients.
Bottom line here is: There in indeed extra noise in the 3-bit
path, amounting to about 15 to 20% over the 8-bit path.
There is (I claim) no asymmetry between RCP and LCP -- we have a couple
of bad antennas/samplers on the RCP side. The origin of the additional
noise is yet to be determined.
_______________________________________________
evlatests mailing list
evlatests at listmgr.cv.nrao.edu
http://listmgr.cv.nrao.edu/mailman/listinfo/evlatests
More information about the evlatests
mailing list