> Frequencies appear to be correct.
Excellent.
> 25R and 26L give perfect (integer) zero amplitudes.
Th Station Boards are claiming all data are invalid for those IFs.
For ea25-A this comes from CRC errors on input C from the deformatter.
For ea26-C the origin is not so obvious, but again the data (in)valid bit
is being set at the Delay Modules.
> 6, 15, and 20 give only noise on both polarizations
ea06 and ea20 have no L band receivers.
ea15 had some circuit fault I believe.
> 26R and 27L give only noise.
ea26-A is mysterious -- the deformatter fibers were misaligned, but that
should just give odd data as seen at C band.
ea27-C's deformatter is either missing or brain-dead.
> Delays are HUGE = +/- 200 nsec. Indeed, with the relatively coarse
> frequency resolution (1 MHz), the phase gradient across the channel
> width (0.45 turns in the worse case!) is sufficient to explain the
> 'closure' errors I reported on yesterday.
Oh good! I started wondering about that just when this e-mail arrived :)