[evlatests] K-Band stability, etc.
Rick Perley
rperley at nrao.edu
Tue Apr 6 18:13:03 EDT 2010
Michael sent me the K-band data that he showed last week, for my
scrutiny. This is what I found ...
The run is about 15 minutes on 3C84, comprising 32 individual scans
of 30 seconds each.
All antennas in the database fringed except 7C -- the cause of which
I believe has been found and repaired.
Curiously, antenna 19 was slow to get back on source following a
reference pointing scan -- an extra 27 seconds.
Antennas 23 and 21 have rather large delay errors -- 10 to 25 nsec.
The rest are all < 10nsec.
Bandpass stability is really quite amazing, with pk-pk changes in
the bandpass, for *most* antennas, being less than 0.4% from the
average. But there are notable exceptions:
* Antenna 2 shows slow large scale (tens of MHz) changes of ~1% in LCP.
* Antenna 6 shows a slow large scale variation, of ~3% in IF B.
* Antenna 11 has 3% upturn or downturns in the edge channels of IF B
(within 5 MHz of the edges).
* Antenna 25 shows the same effect as antenna 11 -- except that the
up- or down- turns track in time, whereas for antenna 11, different
scans were quite different in the edge channel amplitudes.
A number of antennas show slow large scale changes, but (except for
those noted above), the amplitude of the changes is less than 1%.
Amplitude stability is 'pretty good'. In all cases, changes in
amplitude occur equally in all four IFs. There are lots of
'micro-phenomena', nearly always at the 1 to 3% level. Most are likely
changes in gain or pointing -- without the switched cals there is no way
to reliably identify the origin.
Some gain changes are certainly attributable to the antenna beam
wandering away from the source:
* Antenna 13 for the first 5 minutes -- amplitude changes by 6%.
* Antenna 20: Drops by 5 percent over a couple of minutes --
amplitudes were restored following a subsequent reference pointing.
* Antenna 22: Same as antenna 5.
* Antenna 28 shows very strong loss of gain over the three 5
minute observation (between referenced pointing scans) -- gain loss of
20% (amplitude) in one case. Was the pointing especially bad at this time?
* Antenna 5 showed two or three scans with very different
amplitudes.
Phase stability was generally pretty good, but some odd changes were
found, where the phase of an antenna 'jumped' by typically 10 degrees.
In all cases, these jumps occurred on scan boundaries. Some antennas
were particularly bad -- antenna 19 is quite exceptional (and not in a
good way ...), showing 10 degree jumps, up and down, at every single
scan boundary. For most antennas, the jumps occurred only once or twice
over the 15 minutes.