[evlatests] Results from WIDAR 'stress' tests, Saturday afternoon.
Rick Perley
rperley at nrao.edu
Sun Dec 20 14:37:14 EST 2009
Michael took WIDAR data with the standard 'test' setup Saturday
evening. These were run at C and K bands, with the standard setup.
Following the C and K band observations, Michael ran the same script
with the VLA's correlator, providing some very useful comparisons.
The two sources, 0217+738 and 0228+673, were separated in elevation
by 3 degrees. The observations were 40 minute long.
Calibration followed a now-familiar path:
1) Use QUACK to retain only the last 1.0 minutes of every scan. The
scans are 1minute 23 seconds long.
2) Use FRING to find a global (over all 8 contiguous sub-bands)
delay and phase solution for a single 1 minute long piece of data, *not*
selected from the first scan. Use CLCOR to implement that solution for
all sources. (6 hour boxcar).
3) Use CALIB to solve for amplitude and phase for each sub-band and
polarization, using 50 channels (25 MHz) from the middle of each
sub-band. Use CLCOR to interpolate this (using 0217+738 as the
calibrator and 0228+673 as the target), with a 1-minute smooth.
4) Use BPASS to solve for the bandpass, for each source and each
scan (to check for bandpass stability, etc.).
Results are generally very, very good, with a few important
exceptions. Key points:
A) C-band.
1) The first scan has distinctly different phase than all others.
There is a remarkable pattern: the phase change between the first and
second scans has one of 5 different values: 0, 60, 120, 180, and 300
degrees. Both polarization and all subbands for a given antenna jump
by the same amount. The antennas group the following way (antenna 4 was
the reference:
0: antennas 4 and 27
60: 3 and 25
120: 5, 9, and 24
180: 19
300: 2, 8, 15, and 28.
2) Amplitude stability was outstanding, with no identifiable drops
in amplitude, other than antenna 9, which seems to take a couple of
extra seconds to get back on source after a change.
3) 'Zero' records are very rare: typically six for a given antenna
throughout the run (out of about 50,000 total). 'Zero' characteristics are:
- baseline based: if a particular baseline shows a 'zero'
amplitude, there is no tendency for either antenna to have a zero with
any other antenna at that time.
- different between polarizations. The number of zeros is
about the same in R as in L, but the occurences are completely different.
- the same for all channels for a given baseline, subband and
polarization. If one channel is zero, all are.
- different between subbands. Each subband had about the same
number of 'zeros' in the run, but the occurences are different for each
subband.
4) Phase connection between the sources is excellent in general.
Antenna 24 may have a repeatable 5 degree offset betwen sources. Nearly
all antennas connect to with 1 or 2 degrees every time. *** In
particularly, any offset between sources for antenna 8 is limited to
less than 1 degree***. Near the beginning of the run (at 1:18 IAT), a
plausibly atmospheric event caused phase jumps of 10 to 15 degrees
between the two sources, on the longer baselines only. The size of the
jump increases with baseline length.
5) Bandpass stability is outstanding. All subband bandpasses
connect in amplitude and phase. Subband #1, as usual, is useless on its
lower edge. There is something quite wrong with antenna 8, LCP -- the
bandpass shows amazing (but stable) structure of about 3.2 MHz period...
B) K-Band.
1) The initial scan phase jump is only weakly seen. When present,
it is 30 degrees or less.
2) Amplitude stability is similarly outstanding. The same 'late to
get on' behavior with antenna 9 is seen. Antenna 27 has very low
amplitudes in both polarizations and on all subbands, representing a
real SNR loss -- as it did last Friday.
3) There were *no* zero records that I could find.
4) Phase connection is generally good, except on antenna 8, which
has the same 15 degree offset between sources that I found last Friday
(with a different pair of sources, but the same elevation separation).
Antenna 24 also has a clear offset at most times, typically 5 to 10
degrees. Complicating interpretation is evidence of an unusual
atmospheric-like event which passed by during the first 20 minutes or so
of the run -- slowly oscillating phases (period of 8 minutes) which is
clearly a coherent pattern over the entire array. The pk-pk of these
waves (from end of West arm to end of East) is well over 130 degrees!
The phase amplitude clearly scales with distance down the arm. If not
due to the atmosphere, the cause certainly knows the location of the
antennas ... The phenomenon died away by the end of the run.
5) The bandpass characteristics are the same as at C-band.
C) VLA K-band comparison.
Immediately following the WIDAR K-band observation, the same script
was run with the VLA's correlator. Key results are:
1) The 15 degree phase oscillation between the pair of sources seen
at K-band on antenna 8 is also seen, with identical characteristics, in
the VLA correlator data.
2) The smaller phase step seen in antenna 24 is also seen in VLA
correlator data.
3) Phase offsets between the sources for non-WIDAR antennas are very
common in the VLA data: antenna 1 has a 15 degree step. Antennas 10
and 12 have a 10 degree step. Some non-WIDAR antennas have quite poor
phase stability-- or steps -- these are hard to discriminate. The
central point is that I see no evidence that these issues have anything
to do with WIDAR. They occur in equal measure with the VLA correlator.
4) The low sensitivity seen with antenna 27 at K-band with WIDAR is
***not*** seen in the VLA correlator data. This might be interpreted
as something caused by the formatter -- but this antenna gives perfectly
good data at C-band through WIDAR. So the only plausible solution (that
comes to my mind) is in pointing. But this would require that the VLA
K-band observation have referenced pointing done. Looking at the logs
indicates this is not possible -- only one minute passed between the end
of the WIDAR K-band test and the beginning of the VLA corrrelator test...
More information about the evlatests