With the return of WIDAR, Michael took some data on our favorite
point source, 0217+738 on Aug 27.
The source was observed at very low elevation (20 degrees!) -- not a
good idea IMHO. See later notes ...
Observations were taken at X and C bands. I report here on the X
band only, as the Cband data had some sort of translation problems.
All 12 antennas fringed in all four correlations. In general, data
quality is very good. No 'wobbles' are seen. But lots of other, likely
more minor, issues were found. Read on ...
0) The archive claims the observation is an hour long. But only 12
minutes of data were actually present. Michael states that some board
malfunctioned. Why should this completely terminate the observation (so
far as the archive is concerned)?
1) Drop-outs are present. A small number -- << 0.1% of the
visibility amplitudes and phases are identically zero. All four
correlations are affected, independently.
2) The 10-second interval 'drop-downs' are completely absent! (Yea!)
3) Antenna 3, in LCP on sub-bands 3 and 4, has a sharply tilted
bandpass shape: the bandpass slopes sharply downwards with frequency in
subband 4, and oppositely in subband 3. Subbands 1 and 2 look fine.
4) Antenna 8 has a llarge-amplitude sinusoidal bandpass curve in LCP
on all four sub-bands. There is exactly one cycle for this -- with
inverted phase for half the cycle. It looks like a (very) bad lag to
may eyes (or all lags are bad but one).
5) Essentially all antennas show small amplitude and phase
oscillations in both polarizations. The frequency scale is always the
same, corresponding to about 7 meters free-space. Sure sounds like a
reflection between subreflector and feed/receiver to me. Some
antenna-IFs are worse than others -- 9 in LCP is particularly notable.
6) The amplitude gains show curious gain fluctuations which are
different for each antenna, but essentially identical in all subbands
and polarizations for any given antenna. The fluctuation timescale is
fairly short -- 10 to 20 seconds. The fluctuations are not sinusoidal.
I don't think these are 'wobbles'. There is no effect on the phase.
Some antennas show very little of this effect: 1 (W16) and 24 (W14)
show very little effect. Antennas which might be expected to be
shadowed are *much worse*. The amplitude variations for some exceed 10%
!!!!! I suspect shadowing and/or cross-coupling is involved here.
Alternately, we have a serious pointing issue (but I think this much
less likely).
*** I strongly urge that we NOT observe low elevation sources in
these tight configurations. We have enough things to dig out of these
data without worrying about shadowing effects ***
7) Antenna 23 had a large amplitude loss upon return to the
calibrator. (There were only two such observations, due to problem (0),
above). The change in amplitude gain was about 20%.
8) In general, the LCP amplitudes are a little lower than RCP from
the same antenna/subband -- but not always.
9) Delays on the RCP side scatter typically about 10 -- 20 nsec
about the average. The spread is *much* larger on the LCP side -- many
tens of nsec, with some antennas (1 and 2 especially) much further away
from the mean -- over 100 nsec for these two.
I calibrated the data, using a 2-point A&Phase average, integrating
over the ~central half of the bandpass, and applying the mean bandpass
solutions. (The bad LCP subbands from antennas 3 and 8 were flagged out).
Imaging tests showed the following. I had only 7 minutes of good
on-source data -- one cannot expect much with so limited an integration.
1) Imaging with a single, central channel, with 1 to 4 IFs, in
either polarization, produced maps which appear noise-limited. The DR
is a modest 1200.
2) Noise reduced as expected for channel integrations up to about 16
or perhaps 32.
3) Images made with more than 32 channels do not follow the sqrt(BW)
noise expected, and indeed get worse when more ~200 channels are
included. This can (likely) only be caused by a variable bandpass. I
did not further investigate this. The max dynamic range I got was about
11,000. This is very good for a short observation -- I honestly don't
know what to expect here...
4) Application of a closure correction had no discernible effect.
The reported closure levels are very low -- a few times 0.01% in ampl
and phase -- probably the noise limit, as only 7 minutes of data were
available to calculate these.
5) I turned on AIPS' closure reporting statistics -- these are
consistent with the numbers given above. No antennas or subbands 'stood
out' in the statistics. All seems good here.
6) The residual gains, after the 2-point calibration reported above,
showed that the LCP shows considerably greater scatter than RCP -- by
about a factor of 3. Yet the images in LCP are only slightly poorer
than RCP -- most of which can likely be explained by the absence of
antennas 3 and 8. I have no easy explanation for this ...
That's about all I can squeeze out of this, in the 2 hours available
on a Friday afternoon ...
Rick