[evlatests] C-Band Sensitivity Troubles
Rick Perley
rperley at nrao.edu
Wed Jul 16 13:55:31 EDT 2008
I used a few minutes of maintenance time this morning to roughly
calibrate the Tcals for the antennas at C-band, and hence get an
estimate of the system temperatures. This was done by observing Cygnus
A ( a strong source of known flux density), noting the reported rise in
system temperature, and adjusting this by the expected rise.
Cyg A provides about 495 Jy at 6cm. If we make the assumption that
the efficiency of the antenna at this frequency is same for all
antennas, and is equal to 0.55, then the expected rise in Tsys will be
about 50K. Although there will be some variation in antenna
efficiencies, these are most unlikely to be greater than a few percent
-- a far smaller error than the observed spread in Tsys. So I expect we
should be able to calibrate the Tcals to perhaps 5% -- certainly
sufficient to judge whether the observed poor sensitivity at C-band is
due to high Tsys.
The results of this exercise are as expected (sadly). After
correction by this procedure, the 'cold sky' system temperatures for all
antennas (EVLA and VLA) are typically 40 K to 80K, and correlate very
well with the observed sensitivities (as derived from correlator
coefficients, which are independent of the measured Tsys).
Some details:
1) There is no difference in the mean Tsys for EVLA and VLA antennas
-- about 60K.
2) The lowest Tsys values are from EVLA antennas 14, 16, 18, and 4
-- about 35K in both RCP and LCP. Antenna 13 (which also has decent
sensitivity) did not fringe in these tests, and gave a zero degree
increment on Cyg A. . In 2005, Bob Hayward and I measured antenna
13's Tsys (by hot/cold load tests) to be 24K, with an efficiency of
about 0.55. Presuming 13 is similar to the others, the Tsys appears to
have degraded by at least 10K since then -- or the efficiency to have
dropped to about 0.40.
3) All other EVLA antennas have Tsys values higher than 50K --
that's twice the expected (and required) values!!!
4) The three VLA antennas with remarkably good sensitivities have
the lowest system temperatures amongst the VLA antennas -- 20, 22 and 27
all have Tsys values about 45K.
Although not a precise substitute for proper measurement of Tsys
(via correct values of Tcal), these high Tsys values are very unlikely
to be caused by deviant system efficiencies. The strong (but tentative)
indication is that there is something seriously amiss with our C-band
EVLA receivers.
More information about the evlatests
mailing list