[evlatests] Scan-based sensitivity changes, and other odd effects
Rick Perley
rperley at nrao.edu
Wed Jan 23 18:09:25 EST 2008
Review of the antenna-based amplitudes -- from correlation
coefficients -- has shown some very curious (albeit small) effects which
are clearly EVLA based.
1) Scan-based changes in system temperature.
For some antenna-IFs, we are seeing scan-based changes in
sensitivity, which are correlated with changes in system temperature. A
good example is that of antenna 18 at X-band. Through the 2-hour run,
the correlation coefficients are seen to lie in two different states,
separated by about 3% (as seen by an unchanging antenna), or about 6% in
terms of system temperature. The two states always last for the full
scan length -- there are no changes within any 5-minute scan. The two
IFs I had in the run behaved quite differently. Investigation of the
Tsys showed exactly what is needed to explain the loss in coherence --
two different system temperatures, differing by 2 to 3K -- exactly the
magnitude needed!
The only other antenna which shows this clearly at X-band is antenna
26A.
At L-band, however, the effect is more commonly observed: 16
(again, the two IFs have different ups and downs), 19 (especially in A),
21C (a single change of 3%), 23C (a single change of 5%), 24C (multiple
changes of 2%) 26A (similar to 24C). In all cases, a corresponding
change of the appropriate sign was seen in Tsys. However, in these
cases, the magnitude of the L-band system temperature change is not
sufficient to correct the amplitudes completely. For both 21C and 24C,
the post-Tsys correction shows a residual effect of about 1%, in the
sense that the reported rise in Tsys is not large enough to explain the
entire change in sensitivity.
So -- what is causing scan-based small (few K) changes in Tsys? Ken
has wondered if the T304 attenuator setup might do this. If it did, it
would not be seen in the B and D IFs, where we used fixed value of
attenuation. Unfortunately, I cannot judge this, as only IFs A and C
were utilized. On the other hand, I wouldn't have expected a change in
system temperature from minor change in attenuator settings so deep in
the signal path -- the preceding RF and IF gain should be enough to
reduce the additional noise to negligible levels -- shouldn't it?
2) 'Weird Effects' at L-band on antennas 11 and 17.
The correlation coefficients seen on some EVLA antennas are behaving
in a very strange manner which is tightly connected to system
temperature changes. A good example is antenna 11. The correlation
coefficients for both IFs show exactly parallel, essentially random,
changes of about 3%. These changes (unlike those described above) occur
both within scans, and between them. (Note, however, that behavior at
X-band is close to normal). Examination of the Tsys values shows that
there are exactly parallel changes in system temperature -- but in IF
'C' only! Hence, the Tsys-corrected values of visibility look quite
good in IF 'C', but retain the characteristics described above in IF
'A'. Clearly, our measurements of Tsys in IF 'A' at L-band are not valid.
Similarly, antenna 17A shows remarkable changes in correlation, over
a timescale of about one hour (5%), much more strongly seen in IF 'A'
than in 'C'. The measured Tsys values mirror these changes quite well
(and I note the values are 50 to 65K -- two times too high), but the
post-Tsys visibility stability, although much better than that of the
correlation coefficients, still shows residual effects.