[evlatests] EVLA status, Wednesday afternoon
Rick Perley
rperley at nrao.edu
Thu Nov 29 14:19:00 EST 2007
Following maintenance yesterday, the standard continuum stress test
was run to check performance.
Overall, outstanding performance at all bands.
Emmanuel will provide a more detailed review later. In brief:
a) The *only* global VLAl-EVLA phase jump (and probably the only
significant phase jump of any kind, other than those due to atmosphere)
following the P-band observation, which was done at a different bandwidth.
b) The integration error affecting scans prior to a reference
pointing scan, did not occur. It seems that Walter's fix is working.
c) The 'square wave' phase seen in antenna 18 in IFs A and C, is now
gone.
Some band dependent issues are noted below:
P-Band:
Much better now! Antenna 13 still doesn't fringe. Antenna 21 has
its polarizations reversed. Antenna 11 has non-functioning dipole, or a
broken connection between one dipole and the polarizer. Otherwise, all
sensitivities (including antenna 21 -- there's an AIPS task that allows
one to reverse the polarizations, so the data can be analyzed) are
within acceptable bounds.
L-Band:
16B has a high Tsys, resulting in a factor of 2 loss in
sensitivity. As 16A is o.k., this is presumably due to a problem after
the front-end?
A peculiar short event was seen in all IFs of antenna 25: the
corrrelation coefficients all dropped by a factor of four, for about 1
minute. There was no change in Tsys. Easy explanation is that the
antenna was off source -- but how could that be?
C-Band:
A significant drop in correlation was seen on the last scan, for a
number of EVLA antennas: 13C, 16A, 18A, 25A and C, 26 A and C. Therre
was no change in Tsys. The other three scans were normal. This cannot
be due to pointing. No other explanation is offered.
X-Band:
21C was very unstable (variations by a factor of 100 in amplitude)
early in the run. This antenna/IF settled down, and was fine later on.
14C and D (LCP) has high Tsys -- 2X too high.
23C has a sensitivity low by a factor of 2, but Tsys is o.k.
26B has a curious problem: the correlation coefficients are varying
by 10s of percent, on timescales of tens of seconds. But the Tsys looks
stable. However, it is noted that this IF's Tsys did *not* show the
change expected when the antennas slewed onto the source. The other 3
IFs showed the expected change in power as the antennas changed
elevation. This problem does not appear to occur at the other bands,
but a more careful review is warranted.
K-Band:
Antenna 25 sensitivity low by a factor exceeding 10 -- due to high
Tsys (about 300 K)
Q-Band:
Antenna 21 sensitivity low by a factor exceeding 5 on all IFs, but
Tsys o.k. I suspect bad pointing, as the unstable IF noted above at
X-band likely prevented a good reference pointing solution.
Finally:
The EVLA data are littered with the 'Short Shallow Drops'. Although
this is not a critical issue by any means (the loss of sensitivity
should be less than 1%), it would be very nice to diagnose and remove
this long-standing problem. Is there any test we can perform to help
here?