[evlatests] EVLA Holography -- almost working!
Bryan Butler
bbutler at nrao.edu
Mon Mar 5 21:48:20 EST 2007
i am going to look into this tonight or tomorrow. i'm pretty sure i can track
down the cause for 1) and 2).
-bryan
On 3/5/07 17:29, Rick Perley wrote:
> A quick test was done this afternoon to test EVLA Holography. In this
> test, three VLA antennas and two EVLA antennas were declared as
> references. The others executed a standard 11 x 11 raster, with twice
> critical spacing (two points per beam).
>
> Three problems were found -- two of which are fundamental, the other a
> fluke (note: not Fluke!)
>
> A) Fundamental problems.
>
> 1) The EVLA and VLA are not perfectly synchronized. Each array
> is executing the proper pattern, but the EVLA is running about 4 seconds
> behind the VLA in its movements!
>
> 2) The EVLA's reference antennas are being told to go to the first
> offset holography position -- where they spend 10 seconds -- then they
> are told to do what was intended of them in the first place -- and rush
> off to remained fixed on the source.
>
> B) Fluke (NB NOT the Flukes of recent fame) Problem.
>
> All EVLA to VLA baselines were low by about a factor of four for the
> first 20 minutes of the 30 minute test. But the EVLA to EVLA baselines,
> and VLA to VLA baselines were normal, or near normal. Ken has a
> rational explanation for this, involving the fact that the EVLA antennas
> were not being utilized in the run preceding this test, and were not
> properly returned until well into this test. The loss of amplitude was
> because the EVLA test was at 12.5 MHz BW, while the antennas were
> apparently left in a 50 MHz BW. He can explain the curious set of
> circumstances needed to create the effects we see. He is confident this
> is a 'one-of' type of problem. (Right, Ken?)
>
>
> _______________________________________________
> evlatests mailing list
> evlatests at listmgr.cv.nrao.edu
> http://listmgr.cv.nrao.edu/mailman/listinfo/evlatests
More information about the evlatests
mailing list