[evlatests] Phase Jumps and the Flukes
Rick Perley
rperley at nrao.edu
Mon Mar 5 21:44:46 EST 2007
A test was run today to test Barry's hypothesis that changing
frequencies by a value selected so that the Fluke synthesizes are
incremented by a multiple of 10 MHz will result in no phase jumps.
Barry is correct.
Here are the details.
The first 30 minutes of the 60 minute test alternated between two
frqeuencies, 1465 and 1445 MHz for IF #1, and 1385 and 1405 MHz, for IF
#2. These values were selected so that the frequency change could be
set by the Flukes alone (i.e., the L6 settings were made identical).
Specifically:
1465/1385: L6: 3640 and 3560, Flukes: 121.975 and 221.775
1445/1405: L6: 3640 and 3560, Flukes: 101.975 and 241.775.
(Ken reminded me that it is insufficient to change the frequency by
10 MHz, as the Flukes running the B and C IFs increment by half that set
for A and D).
During this first 30 minutes, I alternated back and forth, with a
one minute integration.
The second 30 minutes were spent in the same way, but with a pair of
frequencies that were NOT different in frequency by a multiple of 20
MHz, but which also used the same L6 setting. Specifically
1465/1385 L6: 3650 and 3560 Flukes: 121.975 and 221.775
1450/1390 L6: 3650 and 3540 Flukes: 106.975 and 226.775
For all observations, the BW was 6.25 MHz.
Results:
For the first pair of frequencies (which differed by 20 MHz), all
phases are rock steady. There are no jumps.
For the second pair of frequencies, there are many 'global' phase
jumps -- all of them of the exactly 180 degree variety. The
characteristics of these 180-degree jumps are identical to those
reported earlier (in the five or seven L-band frequency tests:)
All VLA antennas jump identically.
The RCP and LCP sides, for a given IF, jump identically.
The two IFs do not jump in synchronism. For example, for IF1 at
1465 MHz, the VLA was in the 'up' phase statesfor eight of the 13
observations, and in the down (by 180 degrees) state for five. But IF
#2, at 1385 MHz, the 'up' state was seen in four scans, the 'down' in
nine. Meanwhile at the other frequency pair (1385 and 1390 MHz for IF 1
and 2, respectively), the patterns were quite different: IF1: 7 ups
and 6 downs; IF2: 5 ups and 8 downs. There seems to be no correlation
between changes between up and down between IF pairs or frequencies.
It is interesting to note that the 72-degree regular phase
advancement (or retardation) noted in my multiple, cyclic frequency
tests two weeks ago, is entirely absent. The only jumps in this
database are by 180 degrees. Given the manner in which the Flukes were
commanded, it seems pretty clear that the 180 degree family of phase
phenomena originate with the Fluke synthesizers.