[evlatests] System Check today
    Rick Perley 
    rperley at nrao.edu
       
    Wed Apr 25 19:35:56 EDT 2007
    
    
  
    The intention was to compare Modcomp-on and Modcomp-free data, but 
as IDCAF was not operational, I have only the Modcomp-driven data to 
review. 
    Global results:
    The 150 degree phase jumps seen over the past couple of weeks on 
14C, at P, L, and occasionally C-bands, is now gone.  I hope this was a 
result of somebody finding and repairing the problem! 
    The multiple amplitude levels reported recently from antenna 26D are 
also gone.  Same comment given above applies. 
    Antenna 21 worked well now that it points properly.  Fringes were 
found on all IFs for L, X, and K bands.  Still no fringes at P-band, and 
there is no receiver at C and Q bands yet. 
    The last 1 second of data from EVLA antennas, at all bands, was bad, 
but was not caught by the flagger. 
    Antenna 23B and D have 'ramp gains' -- gain changes changing 
linearly, with a 'step' back to the original gain.  Typical timescale is 
40 seconds, amplitude of ramp about 15%.  B and D are not ramping in 
synchronism.  This happens at all bands. 
    Band-Dependent results:
    I filled the data with flags off, and with flags on, to see what the 
system is catching, and not.  In general, the EVLA flagger is overdoing 
it, taking out more data than warranted.  There are some curiosities ...
    1) P-Band.  Antennas 14 and 17 fringing nicely on all IFs.  Nothing 
from 21.  All others have no receivers yet. 
    Most antennas take about 30 seconds to settle down after the 
frequency change (from L-band in this case).  The VLA is catching this 
nicely, so there is little initial bad data to worry about.  The EVLA is 
taking out 70 seconds at the beginning -- quite a bit too much.  In 
addition, 17A was flagged entirely, but in fact the data are good. 
    2) L-Band.  All EVLA antennas worked well, except 23, which was 
stowed to get its Q-band receiver installed. 
      
       The last 20 seconds of data -- from all antennas, was garbage.  
This was missed by both VLA and EVLA flaggers. 
       The antennas take typically 20 seconds to settle down after band 
change.  The VLA's flagger caught the first 12 seconds.  The EVLA took 
out 42 seconds -- much more than required. 
    3) C-Band. 
    The settling time and flagging characteristics are the same as at 
L-band. 
    In addition, the EVLA flagger is letting pass some data very early 
in the scan -- clearly bad data. 
    4) X-Band.
    Only 23 failed to fringe -- it was stowed.  16C had a ramp-like gain 
(similar to 23, described above) -- but only at this band!
    Curiously, at this, and at K and Q bands,  there is essentially zero 
settling time for the antennas.  This is known to the flaggers, which 
flags very little initial data -- correctly.  However, there are curious 
exceptions on the EVLA side:
       Some antenna-IFs taken nearly 30 seconds to 'come into 
equilibrium' -- and this is not caught by the flagger.  Antenna-IFs 
affected are a strange lot:  13A, C, 14A, 16-all IFs, 21B and C, and 
26A, B, and D.  All other antenna-IFs not listed showed no effect -- 
gains were right, from the start.
    But -- the EVLA flagger for some reason tossed out 10 seconds of 
(generally good) data, on all antennas and IFs, about 12 seconds into 
the scan. 
    5) K-Band
    All EVLA antennas worked, except 23 (stowed). 
    Here there is essentially no 'settling' time after band change.  And 
the flaggers did the right thing.  (This band was first in the file). 
Many antennas (both VLA and EVLA) have oddly variable gains in the first 
30 seconds -- these look nothing like the changes commonly seen at the 
beginning of the low-frequency band scans. 
    6) Q-band
    All antennas outfitted worked, except 26.  (Is its receiver gone?) 
    Like K-band, there is virtually no 'settling' time when changing to 
this band (from K-band, in this case).  Nearly no initial flagging was 
needed, and the flaggers flagged essentially nothing at the beginning of 
the scan. 
    However -- all EVLA antennas had 10 seconds of good data flagged, 
after the initial 10 seconds of data (mostly good data) were not flagged. 
   
    
    
More information about the evlatests