EVLA tests of 2005 Nov 23.
Rick Perley reported from data of Nov 18-22 that there appeared
to be closure errors i.e. amplitude & phase variations on the
baseline 14-16, which were less apparent or absent on baselines
14-VLA and 16-VLA. C-band data had a time-variable closure error
on ~4s timescales, X-band was steady.
The goal here was to see if these were real/repeatable. I stuck
to a single point source and did not investigate the elevation-
dependent phase offsets that Rick also reported before.
Continuum data were taken at 3s integration, at X, L, K, C bands.
Line data, 10s integration, were taken at 50, 25, 12.5MHz BW, IF A,
for C, K, L bands. At X-band, IFs A,B,C,D were recorded at each of
the 3 bandwidths.
I report mainly on the continuum data here.
1. Some data may have been lost due to interaction with another
subarray, I have not investigated the details yet; but all IF's
and all bands produced enough good data.
2. No rapid phase wraps or amplitude dropouts were seen. Antennas
14 and 16 were as good as VLA antennas, to 0th order.
Deviations are noted below.
3. There ARE what appear to be closure errors, steady over ~10
minutes at X and L bands, and variable on ~4s at K and C bands.
Differences between the 4 IF's exist at a lower level, I have
glossed over them here but kept plots and notes.
4. The combination of LOs and sidebands in use at the different RF
bands, along with the matrix below, may be useful in diagnosis.
I'll need some help from a hardware expert.
Band amp 14-16 amp 14-VLA amp 16-VLA Comment.
X 0-20% HIGH +_10% +_10% steady
L 0-10% LOW 0-10% LOW 0-10% LOW steady
K 0-10% HIGH = VLA = VLA jitter
C 10-30% LOW 5-10% LOW 10-30% LOW jitter
The ranges are the mean values from different IFs - the jitter
is peak-to-peak deviation from the mean.
The clearest case is C band, where the jitter is invisible on
16-VLA, 14-VLA, and VLA-VLA, but obvious on 14-16. A phase
jitter of about the right amount, 5-10deg, accompanies the
amplitude jitter.
At K band the jitter is worst on 14-16, but still visible and
maybe IF dependent on 14-VLA.
I need to do more on this, e.g. look at TSYS correction? Aliasing?
Frequency dependence across the bandpass? These can be investigated
with the data in hand; I will send an update after I chew through the
spectral line data - a dry turkey, without wild turkey to wash it down.
Cheers, Vivek.