[evlatests] New failure mode!
Rick Perley
rperley at aoc.nrao.edu
Thu Dec 29 10:57:55 EST 2005
I ran detailed sensitivity tests yesterday evening. Detailed
reports will
be made later today, after I've had the time to calibrate the data.
Here I report on a curious new (to me) failure mode.
The observations were made in single-IF spectral line mode -- four
observations for each band -- 20 observations in all on the calibrator,
and 20 more on blank sky.
There were actually 23 calibrator observations -- three band/IFs
were repeated (i.e. mode 1D at X, K, and Q bands was observed
twice -- all other combinations were taken only once).
Of these 23 trials, 6 showed no fringes on the calibrator for
either 14 or 16! Two of these were 'repeat' observations, for which
the other observation showed fringes.
A curious sequence was seen in the failures:
Time(IAT) Band/IF Good?
----------------------------------------------
1:01 X-1D Y
1.04 K-1D Y
1:06 Q-1D Y
1:09 L-1A Y
1:12 C-1A X
1:15 X-1A X
1:18 K-1A Y
1:21 Q-1A Y
1:24 L-1B Y
1:26 C-1B Y
1:29 X-1B Y
1:32 K-1B Y
1:35 Q-1B X
1:38 L-1C X
1:41 C-1C Y
1:43 X-1C Y
1:46 K-1C Y
1:49 Q-1C Y
1:52 L-1D Y
1:54 C-1D Y
1:58 X-1D X
2:01 K-1D X
2:03 Q-1D Y
----------------------------------------
Note that failures (emphasis: both antennas fail!) are in
pairs, separated by 6 successful observations. I claim this
distribution is not random.
The blank sky observations always follow the calibrator,
but I can't tell if the system was working or not on these scans.