[mmaimcal] Change Request No 260: Relaxation of the Band 3 polarization efficiency specification
Al Wootten
awootten at nrao.edu
Mon Sep 27 10:28:13 EDT 2010
Folks,
Although no comment from Science is on the reviewer or approval list for
this change it does affect B3 science. Comments appreciated by the
ridiculously short and impending deadline of 9/30; they may be made on
edm at:
http://edm.alma.cl/forums/alma/dispatch.cgi/changereq/docProfile/101873
260. Relaxation of the Band 3 polarization efficiency specification
CRE Number :
ALMA-40.02.03.00-0662-A-CRE
Description :
Relaxation is sought to the FE Requirement [FEND-40.00.00.00-00226-00 /
AT] which requires that the polarization efficiency of the tertiary
optics system shall exceed 99.5% for all ten bands. The proposed relaxed
requirement (for Band 3 only) is 98% (-17 dB cross polarization in the
integrated sense.) The peak cross polarization for the cold cartridge
alone (without warm optics), [FEND-40.02.03.00-00298-00 / T], is sought
to be relaxed to -19 dB instead of the presently specified -23 dB.
This request is instigated due to two factors: (i) several FE assemblies
have been measured at the integration centers at this level of cross
polarization for Band 3, and this represents the limit of
achievable/measureable performance. (ii) The Band 3 cartridge group is
unable to meet their originally specified cross polarization requirement
(for the “cold cartridge alone”, without warm optics), due to
manufacturability/yield concerns.
Note that, the cross polarization requirements for the Band 3 "cold
cartridge alone" (without warm optics) are specified in terms of peak
cross polarization performance (and not in the integrated cross
polarization performance sense that the completed FE assembly is
eventually evaluated against). This metric was chosen (somewhat
arbitrarily) and meeting or failing this criteria at -23 dB peak cross
polarization level does not guarantee compliance to the formal FE
specifications one way or the other. This metric was chosen because
integrating over the entire beam scan area (again without the warm
optics) was thought to under estimate the cross polarization efficiency.
Statistics collected for the Band 3 cartridges produced thus far
indicate that about 50% of the cartridges have a cross polarization
performance (in the peak sense, on at least one polarization/test
frequency combination) that is poorer that -23 dB with respect to the
peak of the co-polarization pattern.
http://edm.alma.cl/forums/alma/dispatch.cgi/iptfedocs/showFolder/118271/def/def/3719192
CRE Form :
Upload the CRE form:
ALMA-40.02.03.00-0662-A-CRE.pdf
No Zon presence Eduardo Pizarro
Posted on 2010-09-21 12:35
Modified by Eduardo Pizarro on 2010-09-24 06:33
Workflow state changed by Javier Marti on 2010-09-24 07:48
Author:
ksaini
Impact:
Date decision required:
2010-09-30 12:00
Name of supervisor who authorized the submittal of this change request:
Javier Marti (jmarti)
Reviewer List:
No Zon presence Kamaljeet S Saini (ksaini)
No Zon presence Pavel Yagoubov (pyagoubo)
No Zon presence Ravinder Bhatia (rbhatia)
Reviewer List Email Reminder (Days):
3
Approval List:
No Zon presence Dorsey L. Thacker (sthacker)
No Zon presence Gie Han Tan (ghtan)
No Zon presence Javier Marti (jmarti)
No Zon presence Dick Kurz (dkurz)
Approval List Email Reminder (Days):
Clear skies,
Al