[mmaimcal] CRE 173 B7 cross-polarization
Al Wootten
awootten at nrao.edu
Fri Jun 1 15:53:25 EDT 2007
Folks,
I commend to your weekend attention:
173. Band 7 cartridges: Cross-Polarisation
CRE Number :ALMA-40.02.07.00-095-A-CRE
(Attachments on edm)
Description :
The pre-production cartridges generally do not meet specification
FEND-40.02.07.00-00290-00/T. We present in the attached pages an analysis
that fails to account for the amount of cross polarisation that is measured.
While we remain committed to investigate this issue before starting production,
we must at present ask that the cross polarisation specification be set at a
level that can reasonably be met by most cartridges. We ask for a relaxation
of the specification to \u201316 dB, defined as the ratio of couplings to a
calibration load through a wire grid, with the grid wires perpendicular and
parallel to the nominal direction of the E vector.
173.2. Review comments Hans Rudolf
Hans Rudolf 2007-05-04 10:09 Reply
According to the remarks "We (IRAM) remain committed to analysing
the cause for this and possible improvements" there is no need for a CRE,
RfWs for the delivered cartridges should be prepared instead.
However, I do not see an easy way to improve the x-polarisation performance.
Most likely a redisigned optics would result in a major re-design of the band
7 cartridges, thus introducing cost and delay to the project. At this instance
of the project such appears not to be an option.
173.3. Further investigation of Xpol in B7 optics
173.3. Bernard Lazareff 2007-05-31 11:38 Reply
SInce the CRE was submitted, we made dedicated measurements to analyze
the cross-polarization of the B7 optics.
1) we find cross-polarization at a level consistent with what is expected from
the slant geometry of the grid diplexer;
2) a design change that we had considered (rotating the grid to an angle that
produces less Xpol) is not feasible because of the non-trivial relationship
between the grid azimuth and the polarization PA in the telescope focal plane
and at the location of the mixer feedhorns;
3) we find that there is another contribution to cross-polarization from the
IR filter, probably resulting from the material anisotropy induced during the
injection molding process.
See attached memo (go to edm for the memo).
As Hans says, it seems we have no options on this one. We should, however,
specify the compromises this CRE will force on ALMA polarization measurements,
for which B7 is the band of choice. We can always be the lone vote for
a dissenting position.
Clear skies,
Al