some notes from our last discussion of antenna pointing issues:
the antenna pointing spec is currently:
0.7 arcseconds offset pointing accuracy, with pointing calibration
every 15 minutes. blind pointing accuracy is to be 2 arcsec
(this is a new spec which peter told me yesterday).
the 4 antenna motion specs are:
1 - Fast switching
1.5 deg (to within 3 arcsec) in 2 seconds
2 - OTFM (On-The-Fly-Mapping)
0.5 deg/sec - positions read off accurate to 0.7 arcsec, but
antennas do not need to be pointed at the same position
3 - OTFIM (On-The-Fly-Interferometric-Mosaicing)
0.05 deg/sec - antennas point to same position to within 0.7 arcsec
4 - slew to new source
1.5 deg/sec in El, 3 deg/sec in Az
issues we discussed:
- regarding the pointing spec and offset pointing, min has been looking
at this in detail lately. he seemed to think that given 50 12-meter
antennas (a reasonable number, given our latest estimates), you could
easily do a very accurate pointing calibration in less than 1
minute. he was considering using the pointing determination scheme
currently used at OVRO, which is several repeated triangles at the
HP point. several others of us pointed out that the continuous
circles or other schemes may actually be better, and min might look
at these other schemes. min's analysis is similar to, but differs
from, robert lucas' analysis in MMA memo 189.
one thing that might be of use here is to have min actually calculate
the estimated accuracy in the pointing determination if we are
willing to allocate 1 minute out of every 15 minutes to pointing
calibration.
- regarding the fast switching, we briefly discussed how many FS cycles
might be undertaken by the array, which concerns the projected
lifetime of the array. harvey pointed out that FS cycles don't
actually stress the structure very much, since the travel distance is
so small.
- again, on fast switching, we briefly discussed whether we need to
have a spec on the time required to actually reach the final pointing
spec (i.e., rather than 2 seconds to get within 3 arcsec, have the
spec be X seconds to get within 0.7 arcsec).
[peter pointed out to me yesterday that once you get "close" (like
to within 3 arcsec), it is only a couple of antenna oscillations
beyond that to reach the required pointing, so that if the resonant
frequency is 10 Hz, then it is only a couple of tenths of a second
to get the final bit of pointing accuracy. see figures 8-10 of
torben andersen's MMA memo 231.]
- regarding OTFM, a brief discussion resulted in the general agreement
that the spec'ed 0.5 deg/sec rate was sufficient.
- regarding OTFIM, a longer discussion seemed to reach no conclusions.
there is still some confusion among us as to how OTFIM measurements
will actually be undertaken. this is all tied up in how the total
power measurement can be done. if the total power measurement is
done separately, in OTFM mode, then the 0.05 deg/sec is probably
sufficient. if the total power measurement is done at the same time
as the interferometric measurement, then the 0.05 deg/sec is not
fast enough, and you probably need the same 0.5 deg/sec rate as the
OTFM mode has. in the case that the total power measurements are
done in OTFM mode, you still probably want the measurement to have
the same SNR as the interferometric measurements, which implies that
it might take of order twice as long to do a mosaic than if you
collect the total power at the same time as the interferometric data.
- regarding slewing to a new source, brief discussion met mostly with
the telephone equivalent of blank stares, implying to me that the
spec'ed rates are probably OK.