[Gb-ccb] blank sky power spectra / r-factor correction
Brian Mason
bmason at gb.nrao.edu
Mon Feb 27 11:23:17 EST 2006
Attached is the power spectrum from a 5 minute blank-sky drift scan with
the CCB collected on Saturday. Solid line is the power spectrum of the
differenced data, and the dashed line is the power spectrum of the total
power data; CCB input ports are J1-J4 (top, left to right), J5-J8 (next
row down left to right), etc. Axes are log10(frequency) and log10(power
spectra density). Typical 1/f corner frequencies for the better channels
are 10^-0.5 - 10^-0.7 =0.2-0.3 Hz. The two first low frequency channels
(J1 and J5) show similar performance in total power and differenced power.
With the working assumption that the 1/f we see is due to receiver gain
fluctuations multiplying an imperfectly balanced input load, we tried
applying the "r-factor" described in the Planck LFI papers by Seiffert,
Menella et al. The idea is to let
Diff Data = Sig - r * Ref
and adjust r so that the mean difference is zero, which reduces the effect
of gain fluctuations. In effect this is a software gain-balancing made
possible by the fact that we measure total power not just the difference.
Doing this made no appreciable difference in the power spectra. We have
more drift scan and hot/cold load data that we need to analyze, and we
also need to look into the validity of the approximations made in our
simple r-factor correction (such as detectero zero levels). But on the
face of it this would disfavor offsets x imbalance as the cause of our
1/f, and suggest an actual pre-hybrid intrinsically variable offset.
Several channels also have notable refrigerator microphonics (J10, J14;
J4, J8, J12, J16) and most have 60 Hz. These are more prominent than in
hot/cold load power spectra owing to lower overall noise levels.
Brian & Larry W.
--------------------------------------------------------------------
Brian Mason | office: +1(304)456-2338
Associate Scientist | fax: +1(304)456-2229
National Radio Astronomy Observatory | mail: PO Box 2
bmason at gb.nrao.edu | Green Bank, WV 24944
http://www.gb.nrao.edu/~bmason/ |
--------------------------------------------------------------------
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ccb25feb06-alldata-pspec.png
Type: image/png
Size: 64382 bytes
Desc:
URL: <http://listmgr.nrao.edu/pipermail/gb-ccb/attachments/20060227/67c90dd1/attachment.png>
More information about the gb-ccb
mailing list