Hi Rick. Fernando & I got time this Maint day to check out
5 of the 10 EVLA antennas where you reported > 10% cross-pol.
All that we checked out were sequentially at the end of the
west arm (for our convenience). The results were:
ea18:
o Alignment almost perfect (referenced to the quad legs).
o Both dipoles were straight.
o The elements were 180 deg out from "normal". ie: When
the antenna is at EL < 90, the vertical dipole element
that connects to the coax center is on top. (We finally
found one!)
ea22:
o Alignment almost perfect (referenced to the quad legs).
o Both dipoles were straight.
o The elements were oriented "normal" (shield side up
for the V dipole).
ea03:
o Alignment almost perfect (referenced to the quad legs).
o The vertical dipole elements were straight, but the
horizontal were not in alignment--1 side off by maybe
1.5 degrees, while the other side pointed directly at
the center of its quad leg.
o The elements were oriented "normal" (shield side up
for the V dipole).
ea16:
o Alignment almost perfect (referenced to the quad legs).
o Both dipoles were straight.
o The elements were oriented "normal" (shield side up
for the V dipole).
ea19:
o Alignment almost perfect (referenced to the quad legs).
o Both dipoles were *not* straight. For both the H & V
sides, 1 element leg would be off by maybe 1.5 degrees,
while the other side of the same dipole pointed directly
at the center of its quad leg.
o The elements were oriented "normal" (shield side up
for the V dipole).
So, I'm not sure what caused your cross-pol results. Perhaps
the linearity of the dipole elements is showing up, but then
why would ea18, ea22, & ea16 show bad? They were straight.
(As noted by my email "Re: ea26 & ea24 P-band dipole orientation
check" of 20170912, your reference antenna, ea24, also has a
linearity problem with one of the 2 dipoles. Could the reference
antenna's imperfections be transferring to the antennas
under test? Still, the 1/4 wavelength element that was crooked
on ea24 was only off by around 1.5 to 2 degrees. ??.)
-Mert
On 9/22/2017 11:00 AM, Frank Schinzel wrote:
>
>
> On 09/22/2017 10:55 AM, Rick Perley wrote:
>> Some P-band data were taken earlier this week. Frank asked me to
>> check on the quality:
>>
>> I extracted the four scans of 3C147, and used five spectral
>> windows (5 through 9). I find:
>>
>> 1) Antenna 28 was in the barn, and antennas 13 and 14 were both
>> out of service during these observations.
>>
>> 2) Antenna 23 was in a peculiar state -- the 'Y' channel
>> oscillated in delay on a fast (few seconds) timescale. The 'X'
>> channel also had irregular delays, but only for the first two
>> calibrator scans -- it seemed fine for the last two (two hours later).
>>
> This is probably a DTS issue. It has been doing funny things.
>>
>> 3) Requantizers seemed to be working (meaning, they changed level
>> between the second and third calibrator observation).
>>
>> 4) CALIB gains (after applying the RQ correction) showed a wide
>> range of antenna correlated power -- a factor of four between highest
>> and lowest for a given SPW.
>>
>> 5) The 3 MHz bandpass ripple is strong on: 7X (2dB, pk-pk in
>> power), and 17X (2 dB). All others are less than 1dB.
>>
>> 6) With respect to 'golden antenna 24', the cross-polarization is
>> high (above 10%) on: ea03, ea10, ea14, ea16, ea17, ea18, ea19, ea21,
>> ea22, ea27. These are likely to to dipole orientation. A
>> confirmation check would be useful.
>>
> Note ea24 DTS was flaky when the data was taken, if you don't see
> anything crazy in the phases or delays then it might have been ok.
>>
>> Rick
>>
>> _______________________________________________
>> evlatests mailing list
>> evlatests at listmgr.nrao.edu
>> https://listmgr.nrao.edu/mailman/listinfo/evlatests
>
> _______________________________________________
> evlatests mailing list
> evlatests at listmgr.nrao.edu
> https://listmgr.nrao.edu/mailman/listinfo/evlatests
More information about the evlatests
mailing list