[evlatests] Tsys variation with elevation
Rick Perley
rperley at nrao.edu
Wed Mar 1 17:30:08 EST 2017
Observations taken overnight on 10 February were primarily intended
to measure the antenna gain variation with elevation. But these are also
useful to measuring the rise in Tsys with elevation. The Tsys system
worked very well for nearly all antennas and on all bands (notable
exceptions will be identified later). Weather was clear and calm
throughout. I observed two sources rising from 10 to 90 degrees, and
two others setting from 90 to 10 degrees, interleaved. Agreement
between ascending and descending sources was impressively good.
I plotted up the Tsys vs elevation curves. The rise with
decreasing elevation is of course quite obvious -- this is due to a
combination of increasing atmospheric emission and ground spillover. It
is harder than you may think to separate these... (see the old memos by
Bob Hayward and me on this topic).
Below is a short table, giving the typical *increase* in system
temperature at elevations of 10, 20 and 30 degrees compared to 90
degrees. The ranges given mostly reflect the uncertainty in the Tcal
values, except at Q-band, where it is due to the rapidly increasing
opacity at the upper end of the band. (Observations at Q-band were made
at 42 and 48 GHz).
Band T10 T20 T30
------------------------------------------------------
L 12 -- 15 5 -- 6 2 -- 3
S 18 -- 25 5 -- 9 2 -- 4
C 15 -- 18 4 -- 5 2 -- 3
X 10 -- 15 3 -- 5 2 -- 3
Ku 17 -- 20 6 -- 7 3 -- 4
K 30 -- 40 11 -- 14 5 -- 8
Ka 35 -- 45 10 -- 16 4 -- 9
Q 60 -- 120 25 -- 60 12 -- 35
----------------------------------------------------------
A few clearly erroneous values are noted:
1) ea08 at C-band claims a system temperature of ~15K. It should be
nearly twice that -- the Tcal values must be wrong. (Also, the
atmospheric rise is about half of what it should be).
2) ea26 at Ku-band claims a system temperature of 70K, and its
temperature rise with elevation is about twice the other antennas.
Thus, the Tcal values here are incorrect.
3) ea03 at Q-band claims a system temperature of about 400K. Since the
atmospheric component here is about right, the problem must lie with
the receiver.