[evlatests] Focus motion failures on ea05 and ea27.
Rick Perley
rperley at nrao.edu
Mon Mar 28 16:45:26 EDT 2016
The data taken Saturday night have revealed an occasional failure in
focus motion on two antennas.
In this run, I did X-band referenced pointing every 50 minutes. The
subsequent observations were done at Ka-band.
For ea05, on six out of the eight referenced pointing observations, the
scan immediately following the X-band referenced pointing was flagged
out. Investigation showed that the flag was set by an incorrect
subreflector position. The amplitude of that scan shows it down by
about 30% in power -- about right for a 7 mm subreflector error (which
is the difference between the X band and Ka band position). The flag
table also reveals that the subreflector position error was cleared
about 4 seconds into the next following observation -- about right for
the required motion.
The conclusion seems clear: the command to move the subreflector,
following the X-band observation, was either not sent, or not received
by ea05, for six out of eight situations.
There was only one other occasion of this -- ea27, which failed the
same way, but on only one occurrence.
Although not common, this problem is not rare -- in my earlier
'beam' holography done last month, this characteristic was seen in most
of the high frequency runs, on at least one antenna. I can get better
statistics (names, numbers ...) if it is useful to aid in the diagnosis
of this failure mode.
Rick
More information about the evlatests