[evlatests] Strong on-linearity at 335 MHz (P-band)
Rick Perley
rperley at nrao.edu
Fri Sep 27 18:16:31 EDT 2013
I have reduced the data taken in the September run on the major
low-frequency calibrators. These data were taken during the
reconfiguration time, and total 12 hours, over a 14-hour period.
I'll discuss the 'PDif' compression issues at another time. Here I
report on the flux transfer results -- these are independent of the PDif
compression problems.
First -- the good news:
The ratios amongst the major interferometric calibrators: 3C196,
3C286, 3C295, and 3C123, are the same (to within ~1%) with the new
system as with the old (P-band) system. Below I give the current ratios
and the mean ratios determined during the many (about a dozen) 'flux
density' runs.
Source Pair Old New
-------------------------------------------------------------
286/295 0.430 0.430
196/295 0.776 0.769
196/123 0.322 0.318
196/286 1.806 1.787
123/286 5.62 5.61
123/295 2.41 2.42
-------------------------------------------------------------
Using the Scaife-Heald flux density of 3C196 as the reference (45.86
Jy), I get the following results for the other calibrators (not
including the FourBigSources, discussed below). All these are for 335
MHz.
3C48 43.1
3C123 144
3C138 16.64
3C147 52.76
3C196 45.9
3C286 25.7
3C295 59.7
3C380 40.4
B) The Bad News
Sadly, the good results listed above do not extend to the four
'absolute' calibrators: 3C144 (Taurus A), 3C274 (Virgo A), 3C405
(Cygnus A), and 3C461 (Cassiopeia A).
Using the values listed above for the (weak) interferometric
calibrators, I then determined the flux densities of these four sources.
Below is a table, showing the 'expected' and 'derived' values.
Source Baars VLA Ratio
---------------------------------------------------------
3C144 ~1235 1080 0.87
3C274 730 630 0.86
3C405 5938 4200 0.71
3C461 6305 3910 0.62
----------------------------------------------------------
We can ignore 3C461 -- Cas A is known to have a secular decrease, so
the very low ratio for this source is definitely too low.
But we can't ignore the others. A deficit of 5 to 10% might be
easily explained. But not a 30% deficit (in the case of Cygnus A).
There is no way the pre-1980s work could be wrong by this factor! The
VLA is clearly seriously saturating -- somewhere -- when we observe
these stronger objects.
This is definitely a global problem -- all antennas are doing it.
As noted above, for 3C144 and 3C274, the various antennas agree quite
well in their gains -- so they have all dropped by about the same
fraction, typically 10 -- 15%.
For Cygnus A, although there is great scatter in the calibrated
visibilities, no antenna pair (baseline) gives a visibility even close
to what is correct.
Fortunately, a clear explanation of this is not hard to find.
In comparing the gains needed to put the fluxes of those four
sources to the right level to the actual power levels seen in the
station board (i.e., correlating the CALIB gains against the PSum
values), it's dead-easy to see what the problem is:
*** We are running the power levels far too high ***
Cas and Cyg typically triple to quadruple the total system power.
The recorded PSum values during the experiment, while on either of these
sources, were typically 60 counts. Due to the troubles in the
'set-and-remember' procedures that I've reported on earlier, we actually
had digital power levels varying between 10 and 120 counts. (!). In
comparing the gain calibration values against the PSum power, I find
that the only instance in which the gains, on Cyg A, were correct, was
when the power level was near 10 counts (one antenna, in one
polarization, for a single 2-hour block). (This was ea03, in 'R'
polarization). Even when we were near the desired 14 counts, it appears
that a significant gain correction (more than unity) was required.
I believe it is critically important to follow up on this was an
on-off Cygnus A experiment, where we set the on-Cygnus A power to 14
counts -- or less -- then compare this to a nearby standard calibrator
(3C380 will do well here).