[evlatests] System Gain sensitivity on temperature
Rick Perley
rperley at nrao.edu
Thu May 9 18:01:51 EDT 2013
The date for the 30-hour flux densities run was partially selected
to occur on a day with a large temperature change. The change from
night to day on May 02/03 was a bit more than 40 degrees F (from 20 to
over 60F). This enables a close check on our system gain sensitivity on
temperature. Two antennas were outfitted with temperature sensors (I
think on the L and Ku band horns) to enable better determination of the
gain sensitivity on temperature.
The switched power gives us an good indication of overall gain
changes -- they are very significant at some bands. Here is a short table:
L-band: ~2%
S-band: 3 -- 5%
C-band: 5 -- 9%
X-band: 6 -- 10%
Ku-band 25%
K-band 25%
Ka-band 12 -- 25%
Q-band: 20 -- 35%
Notes:
1) The gain variations are almost certainly due to the 'heat pipe'
effect (pointed out by Bob Hayward a couple years ago) -- the horns
stick out into the night air, and are good heat conductors. The
temperatures of the post-amps are more affected by the horn temperature
than the air temperature.
2) For all bands, the maximum gain is at the minimum temperature.
This is as expected for a temperature origin.
3) The gain change should be corrected for by application of the
switched power -- *** provided that these devices outputs are either
insensitive to temperature change, or are maintained at a constant
temperature. Application of the PDif values to the visibility data, and
careful calibration, should enable us to determine by how much the
switched power outputs are varying over temperature/time. Analysis of
the Jan 2012 data (the last flux densities run) indicated a ~2% residual
at X and Ku bands -- the only bands whose pointing residuals are small
enough to enable extraction of such a small effect.