[evlatests] Slow Wobbles -- back with a vengeance!
Brent Carlson
brent.carlson at nrc-cnrc.gc.ca
Thu Jan 17 15:20:03 EST 2013
Rick,
On 17/01/2013 12:06 PM, Rick Perley wrote:
> Brent:
>
> The amplification of the effect at the bottom edge of subband #1 is
> likely from the bandpass correction. The attenuation is caused by an
> anti-aliasing filter within the T304. This suggests that the effect
> is added in after this filter. But then this would seem to contradict
> the indication that the effect is proportional to the correlated
> power, which (I think) would cause the wobble amplitudes to decrease
> after that filter. On the other hand, if the effect is related to
> some complicated aliasing from some non-linear process, then the
> filter effect would be avoided.
> We do see, very faintly, wobbles at other IFs. However, in nearly
> all cases, these are only seen in the two or three channels on each
> side of the subband filter 'notches'.
This would indicate that it is a correlator effect since, presumably,
the sub-band band-pass correction would amplify the effect in a similar
manner as you indicated for the T304. Curious, though, why it doesn't
show up or is much weaker in some sub-bands and is weaker in ~1/2 of
sub-band 2...
Setting up non-beating corr chip integration times would be the first
test I'd run...
--Brent
> I hope these clues help ...
>
> Rick
>
> Brent Carlson wrote:
>> Hi Rick,
>>
>> It is strange that even though it shows up in sub-band 2, it is not
>> uniform across all channels in that sub-band.
>>
>> A possibility is that in A config, with a few Hz of earth-rotation
>> phase rate, that the corr chip dump/blanking time is beating with the
>> net phase rate in the correlator...an effect that is known to cause
>> issues. I don't believe that corr chip integration times are set to
>> mitigate this effect as pointed out in NRC-EVLA Memo #032. I also
>> don't know if 3C84 for these observations is located such that very
>> low delay rate effects, if not taken out by post-correlation
>> "discrete-step delay correction", can become evident as amplitude and
>> phase oscillations--although you'd think this would show up the same
>> in all sub-bands and be increasingly worse moving away from the
>> center of a sub-band, and normally an acceleration signature is seen
>> as well.
>>
>> I also note that it seems worst at the lowest edge of sub-band 1,
>> which turns out to be the highest I/F frequency region feeding into
>> the sampler (i.e. since it is Nyquist zone 2, with resulting
>> frequency sense flipping, unless the plots are already corrected for
>> this), and is normally caused by sampler clock cross-talk into the
>> sampler. Don't know what this means exactly, if anything.
>>
>> --Brent
>>
>> On 17/01/2013 10:57 AM, Rick Perley wrote:
>>> I reported, a couple weeks ago, that we had baseline-dependent
>>> slow (periods of seconds to minutes) oscillations in amplitude and
>>> phase on some baselines. These showed up in Ku-band data on 3C84 (a
>>> strong, unresolved object) in A-configuration. The magnitude of the
>>> effect was modest -- a few percent and a few degrees. The effect
>>> was seen only on IF 'C', and it was clear that it was a non-closing
>>> effect (i.e., cannot be removed by antenna-based calibration). A
>>> subsequent test, done at the same band, with the same source, with
>>> the same setup, failed to find these 'wobbles'.
>>> But ... They're back! About ten days ago, we observed 3C84
>>> at Ku and K bands (alternating), while still in A configuration.
>>> The duration of this test was 6 hours, and the goal is to find our
>>> 'ultimate' dynamic range -- and to uncover any remaining subtle
>>> effects produced by our system. Overall, the data quality is
>>> simply amazing. Closure levels (except for what I'm about to report
>>> on below) are effectively unmeasureable -- likely less than 0.01%.
>>> But in reviewing the data carefully, the 'slow wobble' problem
>>> found earlier is present throughout the entire new dataset. But
>>> since we have two frequencies in this long dataset, some new and
>>> hopefully useful information has been found.
>>> 1) The strong slow periodicity is seen in both K and Ku bands for
>>> each baseline on which the effect can be seen. The periodicity
>>> seems unrelated to the baseline coordinates. (This has yet to be
>>> proven). Some antennas are more likely to show wobbles than
>>> others. There is no spatial relationship between antenna pairs
>>> showing the effect. There is no apparent relationship with antenna
>>> number or f-shift. 2) For both, the wobbles are only seen in IF
>>> 'C'. One can make a case for them being very faintly visible in the
>>> others IFs, but the effect is at least an order of magnitude lower.
>>> 3) The wobbles are seen identically in amplitude and phase, with
>>> the same period. Viewed in the complex plane, the 'wobble vector'
>>> circles about the (stationary) visibility vector.
>>> 4) The effect is not visible on a secondary source that was
>>> observed. This source has 1/15 of the flux of 3C84 -- from this we
>>> conclude that the wobble effect is multiplicative, not additive, and
>>> is proportional to the correlated power -- not the total system power.
>>> 5) The effect is strongest by far in subband #2 (#1 in
>>> CASA-speak). It is also easily visible in subband #1, and faintly
>>> visible in subbands 5 and 6. The wobbles are contiguous between
>>> subbands 1 and 2 -- which Ken says is evidence that the baseline
>>> boards are not to blame. (See the attached figures).
>>> 6) The period of the wobbles is shorter at K band than Ku band,
>>> roughly at the ratio of the sky frequencies. For the baseline 2 x
>>> 23 (see attached figures), the wobble period at K-band is 20
>>> seconds. At Ku-band, the period is 30 seconds. The periodicity of
>>> the wobbles is different on all baselines, varying from a few
>>> seconds to a few tens of seconds.
>>> 7) The amplitude of the wobbles is much higher in K-band than in
>>> Ku-band. At K-band, the effect is about 5% of the source flux, and
>>> 3.5 degrees in phase. (pk-pk). At Ku-band, the effect is about
>>> half of this.
>>> The latter two characteristics make us wonder what happens at Ka
>>> and Q bands. A short test this evening -- if there is time
>>> available -- might help here.
>>> I have attached four plots to show the effect -- amplitude and
>>> phase for Ku and K bands. All plots are of the baseline which shows
>>> the effect most strongly: 2 x 23.
>>> About the only thing I'm willing to conclude from this is that
>>> the effect is not caused by an external agent, such as RFI.
>>> Speculations as to origin, or suggestions for further testing,
>>> are welcome!
>>
--
Brent R. Carlson
Brent.Carlson at nrc-cnrc.gc.ca
Tel: 250-497-2346 | Fax: (250) 497-2355
Design Engineer | Ingenieur Concepteur
National Research Council Canada | Conseil national de recherches Canada
Dominion Radio Astrophysical Obs. | Observatoire federal de radioastrophysique
P.O. Box 248, 717 White Lake Rd | C.P. 248, 717 Rue White Lake
Penticton, BC, Canada V2A 6K3 | Penticton, (C.-B.), Canada V2A 6K3
Government of Canada | Gouvernement du Canada
"When and where humans are involved, mistakes inevitably happen"
More information about the evlatests
mailing list