See below
Rick Perley wrote:
> A test (for PDif compression) reveals some clearly incorrect Tcal
> values at Sband:
>
> ea01: Tcal appears to be 10X too high.
Yes indeed; the soida file has an extra column containing Tcal*10
which was loaded instead of the proper column. Corrected.
> ea02 and ea12: 2X too low
The S band receiver in ea02 does not identify itself in monitor
data. MIB is pingable but appears unresponsive. Will fix Tcals
when receiver identifies itself.
ea12 - those are the numbers in the soida files in the 'Current' area,
and are about what they should be.
> ea07: 4X too high
These are the numbers in the soida files in the 'Current' area.
There are other soida files with numbers a factor of two (not four)
lower.
> ea08: 2X too high
These are the numbers in the soida files in the 'Current' area,
and are about what they should be.