[evlatests] Major PDif problem on 3-bit paths
Rick Perley
rperley at nrao.edu
Mon Nov 14 12:36:20 EST 2011
The test run Friday night with the 3-bit system has revealed a
serious issue with the switched power system. The problem is the same
as we've seen with 8-bit, but much, much more severe.
In reviewing the data, I looked at the values of PDif, PSum, and the
Tsys values generated by these first two. Right away a major issue was
seen: For all of the 3-bit equipped antennas, the PDif values are
strongly modified by the total power. The effect is not small. 3C84
adds about 3K to the system temperature (10% of cold sky). When we move
the antennas from 3C84 to cold sky (a 10% drop in total power) , the
values of PDif usually *increase* -- by up to 5%. This is huge, and we
know it's wrong, since the derived value of Tant from observations of
3C84 lead to an efficiency exceeding 75%. An impossibility.
Each antenna and polarization behaves differently. (I haven't yet
checked if the subbands behave differently too). And -- significantly
-- the behavior is quite different between A1 and A2, and between C1 and
C2. Since these pairs come from the same receivers, and go through the
same IF electronics *until the T304* -- the differences noted above must
be generated either within the T304 or the digital system which follows.
Not all antennas show PDif compression when going onto 3C84. ea14
shows expansion -- the PDif values get larger when on the strong source.
On the 8-bit side, all is good, except for ea10 on LCP, where a
faint PDif expansion (by 0.5% only) is visible.
The data taken will not allow us to determine whether the
visibilities are affected by the problem -- i.e., is the problem one
which affects the actual data, or is it solely in the monitor data.
Experience with this problem on the 8-bit side (as shown at C-band)
leads me to predict that we will not see an issue in the visibility
data. But a test is needed to show if this prediction holds.
Since we were not successful in the Friday evening test to show any
temperature-dependence effect for polarization (because the night was
cloudy and the temperatures hardly changed), I propose repeating the
Friday test, and adding in a third source, with flux density ~ 1% of
3C84 (easy to do). We'll pick a good-weather night, with maximum
temperature swing.
Never a dull moment here ... :)