After all those rosy description of PDif working at C through Q
bands, I give a dose of reality for Sband.
Things are not so simple here.
1) Bad PDif values.
- ea6 in LCP has rapidly varying (timescales of seconds) PDif
values which are not seen in the visibilities.
- ea16A also varies in PDif, but not in the visibilities.
- ea07 in LCP has very low PDif amplitudes (1/10 of normal),
and 7C in particular has very high noise.
- ea12C also has high noise.
- ea09A is dominated by noise.
- ea27C and D are very noisy.
*** note that none of these gives a zero PDif -- all have
reasonable values, on average. But there is no evidence that these
values relate to system gain!
2) Changes in PDif not related to gain.
- For a number of antenna-IFs, the PDif varies over time in a
way that makes it look likely it is responding to elevation or
temperature. In particular, ea15, on all IFs, shows a 15% rise in PDif
over a couple of hours while the antenna was rising from 8 through 30
degrees elevation. This change is *not* seen in the visibilities, so
must represent a change in something else, such as the noise diode
power. (The effect is by far the largest in this antenna).
- ea28, in LCP, also shows this effect, but at the 10% levels.
- A significant number of antennas show the effect at lower
levels.
3) Troubles in Tsys or Tant.
- The temporal variability described in Section (1) above
shows up in the estimates of system temperature -- these variations are
thus all (or mostly) in the values of PDif, and not in PSum.
- The claimed Tsys for ea07 in LCP exceeds 200K. This might
be right.
- Due to the variability, or other effects, the expected ~1K
increase in Tsys due to observing 3C84 is completely invisible in ea07 C
and D, ea09A, ea12C, and ea27 C and D.
These are significant issues, which need some investigation.