[evlatests] A Most Bizarre Event
Dan Mertely
dmertely at nrao.edu
Wed Mar 9 18:47:09 EST 2011
Hi Rick. I just checked the W8monitor plot for yesterday, around
6:30 PM and didn't see anything out of the ordinary. It seems
unlikely that ea05 RCP saw it, otherwise it should have shown up on
the plot. (attached) -Mert
Rick Perley wrote:
> The short L-band observation taken last night (around 7PM) was
> affected by an event of unknown origin which greatly modified the
> switched power table for one scan.
>
> There are four scans, two of the calibrator and two of a nearby
> (offset 1 degree or so) blank field.
>
> The first three scans are completely normal. All switched power and
> visibility values appear normal.
> The last scan (duration ~ 2 minutes) was disrupted by some sort of
> event which greatly modified all switched power values, and the
> visibilities.
>
> The major characteristics are:
>
> 1) All antenna were affected equally.
> 2) All subbands were affected, but not in the same way.
> 3) Effects were seen in both PDif and PSum, but they are not the same.
> 4) The effect caused the monitored power to change *smoothly* during
> the 2 minute scan. The preceding scan (a few seconds earlier and only 1
> degree away) is fine.
> 5) The effect is *much* stronger in RCP. For some subbands, there
> is no effect at all in LCP.
>
> Most remarkable is how the PDif and PSum were affected differently,
> for different subbands. In all cases, the PDif dropped, typically by a
> factor of four. For one subband (4), PDif is actually negative! But
> PSum dropped for 12 subbands, and rose for the other four. The four
> subbands with an apparent power increase are: 4, 9, 10, and 15. All
> antennas apparently behave in the same way.
> It's hard to imagine how the switched power can drop by a factor of
> 4 while the summed power can rise by about the same factor.
>
> I've looked at the cross-power spectra, and there is certainly
> strong evidence of an unusual event, seen in some subbands, occuring
> during this time. Some spectra are simply blank (presumably, all
> integer zero), and others are clearly messed up. But other subbands
> look fine during this time -- but evidently are not, as judged by the
> system power monitors.
>
> The IAT time for this event was from 1:38 through 1:42 (or, 6:38 to
> 6:42, PM, local time). It would be useful for someone to look at other
> monitors to see if something very odd was occurring then.
>
>
> _______________________________________________
> evlatests mailing list
> evlatests at listmgr.cv.nrao.edu
> http://listmgr.cv.nrao.edu/mailman/listinfo/evlatests
-------------- next part --------------
A non-text attachment was scrubbed...
Name: SAHP_1500_20110308.pdf
Type: application/pdf
Size: 358658 bytes
Desc: not available
URL: <http://listmgr.nrao.edu/pipermail/evlatests/attachments/20110309/68a3012c/attachment.pdf>
More information about the evlatests